Dear friends
We have a problem with Aloka Echo SSD-500
The problem that left half side from the screen shows abnormal white horizontal lines or we can say a white screen(narrow spaces between the lines) and that make the photo unreadable, but the right half of the picture is ok
then new problem happened and added, that the system can't sense that the probe is connected(no probe)
Am sure that not from the monitor because when we disconnect the probe the problem disappear
an y body can help
regards
Type your question or comment here...
There is an area of memory that corresponds exactly to the image formed by the transducer , which has a number of well defined circuits, there have the problem , or a circuit, or are the copper lines , no more ; try to find the flaw in reverse order to how you would read this message
That is , from right to left and from bottom to top (in the memory bank ) , it appears that well achieved it go looking for the fault that your time in the display (from left to right and top to bottom) , as banks are in parallel, you have no other go reviewing one by one. You have success.
(4)
M.O.A.
a problem with Aloka Echo SSD-500
July 01, 2015 08:42
Dear friends
We have a problem with Aloka Echo SSD-500
The problem that left half side from the screen shows abnormal white horizontal lines or we can say a white screen(narrow spaces between the lines) and that make the photo unreadable, but the right half of the picture is ok
then new problem happened and added, that the system can't sense that the probe is connected(no probe)
Am sure that not from the monitor because when we disconnect the probe the problem disappear
an y body can help
regards
Type your question or comment here...
to rate and post a comment
Roberto Rojas Zavala
re: a problem with Aloka Echo SSD-500
July 01, 2015 11:30
There is an area of memory that corresponds exactly to the image formed by the transducer , which has a number of well defined circuits, there have the problem , or a circuit, or are the copper lines , no more ; try to find the flaw in reverse order to how you would read this message
That is , from right to left and from bottom to top (in the memory bank ) , it appears that well achieved it go looking for the fault that your time in the display (from left to right and top to bottom) , as banks are in parallel, you have no other go reviewing one by one. You have success.
to rate and post a comment
(4)
M.O.A.
re: re: a problem with Aloka Echo SSD-500
July 02, 2015 05:21
Many thanks for your replay
Normally these memory on which board are located exactly
Then what about the sense of no probe??
regards
to rate and post a comment