SEARCH
Current Location:
>
>
> AGFA CR85-X

AGFA CR85-X

(12)
(1)
(4)

Mike Rodemers

AGFA CR85-X

August 20, 2013 04:42

Does anyone know what error code 226C4 means upon power up of CR 85-X. Please advise or if you have service manual with error codes for this unit

Log inor Register

to rate and post a comment

(53)
(13)

Rebel Cook

re: AGFA CR85-X

August 20, 2013 07:10

Power supply test of stepper node belt failed. Check the cables and connectors, fuses, power supply, safety interlocks.

Log inor Register

to rate and post a comment

(34)
(4)
(6)

Luis Diaz

re: AGFA CR85-X

August 20, 2013 07:10

Hello Mike,
The next explications is referent your error in the CR85-X...

226C4
E_IBF_TST_SNBELT_POWER
Reason : Input buffer module: Power supply test of stepper node Belt failed.
Meaning : No Power supply on stepper node detected.
- power supply not connected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.

Good luck!
Bye

1 person liked this
Log inor Register

to rate and post a comment

Altima Diagnostic Imaging Solutions - Call (469) 644-7063

Providing dependable, competitive service for Siemens, Philips, and Toshiba MRI and CT. Experts in Oxford Magnet diagnosis and services. Altima offers ramp and shim services for Siemens TIM and many other MRI systems.

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2015 DOTmed.com, Inc.
ALL RIGHTS RESERVED