I have an error code 4202 on digitizer cr 30-X
who can help. that means the Digitizer has can not give the Layer into the cassett

Jamal

Thanks

1 person liked this
Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: error code 4202 on digitizer cr 30-X

November 26, 2014 08:18

Description:
Start-up: an IP was found in the Digitizer, but no cassette. The IP will be transported to the cassette slot as soon as you confirm. Please confirm and withdraw the IP.
Meaning: An IP was found in the Digitizer, but no cassette during start-up.
Explanation:
Reason: IP-jam occured before or manipulation by hand.
Cure:The IP will be transported to the cassette slot as soon as you confirm. Please confirm and withdraw the IP.