error code 4202 on digitizer cr 30-X
November 25, 2014 03:53
I have an error code 4202 on digitizer cr 30-X
who can help. that means the Digitizer has can not give the Layer into the cassett
Jamal
Thanks
to rate and post a comment
(39)
(1)
Konstantin Kuznetsov
re: error code 4202 on digitizer cr 30-X
November 26, 2014 08:18
Description:
Start-up: an IP was found in the Digitizer, but no cassette. The IP will be transported to the cassette slot as soon as you confirm. Please confirm and withdraw the IP.
Meaning: An IP was found in the Digitizer, but no cassette during start-up.
Explanation:
Reason: IP-jam occured before or manipulation by hand.
Cure:The IP will be transported to the cassette slot as soon as you confirm. Please confirm and withdraw the IP.