faulty hitachi eub-555(no probe message on screen)

(10)
(3)

Okunade Joseph

faulty hitachi eub-555(no probe message on screen)

May 22, 2015 07:54

Can someone help me with my faulty hitachi eub-555. The machine indicates no probe even when I connected a known working probe with it. I have checked for a missing voltage and all seems to be present especially when no probe is connected. But when I connect a probe, the voltages +125v & -125v will reduce to something around +-48v. And the board on which the voltages are was tested on another machine and work very well. Please, what else can I do to make the machine see the probe?
I will also appreciate it if I can get the service manual for this machine. My email is: roupecrowwn2006@yahoo.com, Thanks.