If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Richard Hart
Need Toshiba Artida Service manual
April 18, 2013 09:15
I need a service manual to find out what a number 1 code means on the back of the machine during boot up.
to rate and post a comment
(18)
Uldis Verdins
re: Need Toshiba Artida Service manual
April 19, 2013 09:31
This LED lights green when the 5-V power is supplied to the system startup circuit in the CPU ASSY (the breaker on the rear of the system is ON).
to rate and post a comment
Richard Hart
re: re: Need Toshiba Artida Service manual
April 19, 2013 10:41
The unit has AC and 1, 2 lights green when the breaker is on. On the other side a green number 1 shows up. Any idea what to check?
to rate and post a comment
Junaid Yaqub ali
re: re: re: Need Toshiba Artida Service manual
October 08, 2015 03:15
Hi dear I need the service manual can you send me please at yaqub.ali7@gmail.com
to rate and post a comment
(18)
Uldis Verdins
re: Need Toshiba Artida Service manual
April 19, 2013 09:31
One more detail:
If the AC, 1, or 2 LED is not lit green even though the breaker on the rear of the system is ON and AC input is supplied, the ACDC-UNIT may be defective
to rate and post a comment
Junaid Yaqub ali
re: re: Need Toshiba Artida Service manual
October 08, 2015 03:15
Hi dear I need the service manual can you send me please at yaqub.ali7@gmail.com
to rate and post a comment
(10)
Vu Toan
re: Need Toshiba Artida Service manual
October 14, 2015 07:05
I can help you. Pls contact vutoan_medical06@yahoo.com
to rate and post a comment
(10)
Vu Toan
re: Need Toshiba Artida Service manual
October 14, 2015 07:05
I can help you. Pls contact vutoan_medical06@yahoo.com
to rate and post a comment