Problem in a GE logic alpha 200

Tags: Ultrasound, GE

(53)
(9)

EFREN FRANCO

Problem in a GE logic alpha 200

May 21, 2013 02:03

When turn on the US only appears the image with no data in the screen (study, ID, Patient, Etc) only has the date and time but both are wrong. I tried to change the settings in the menu but doesn`t change and the keyboard stop responding at this point. Does anyone have an idea about this problem?

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: Problem in a GE logic alpha 200

May 21, 2013 07:09

Can you please send a picture of image on Monitor? maybe the CPU board is bad

Log inor Register

to rate and post a comment

(8)

(9)

Bill Burnett

re: Problem in a GE logic alpha 200

May 21, 2013 07:10

System has "lost it's memory" Needs to be re-booted!

Log inor Register

to rate and post a comment