The monitor on the Sequoia will not go on.When I put another monitor on the machine, it starts loading programs but stops and doesn't load anymore.Does anyone have a solution to this problem?
Thank you,
Dinu
Dinu, hi!
I. You can try: press MULTIHERTZ and V KEY
during power on = you may have more informations
about boot sequence untill loading stops.
II. The most likely causes of a Sequoia system
hang:
1. SSP failure on the RDP board
2. Missing Clock signals from RX. (Secondary
clocks from CN)
3. Faulty Hard Drive
4. Misconfiguration
5 Corrupt data
6. SCSI problem
7. RI or AUX Preamp not seated correctly
I hope this will help you.
Marko
P.S. CAUTION!
I. Do not replace the following
combinations of FRUs at the same time:
1. Hard drive and RDP board
2. Hard drive and PIC board
II. Before replacing an FRU, it is recommended that the system be in
the service enabled mode.
Hi Marko Shevchuk
Could you help me ? My sequoia 512 have applications erro message when booting and it stop boot after that
Please show me how to solve this application erro
My email drthienphuoc@yahoo.com.vn
re: re: re: re: re: Re: Problem with Acuson Sequoia 512
August 25, 2016 04:36
We are getting the system error that the MO Disk was not ready after just booting the system up while it still has the boot up gray screen with picture of the tree and will not go past this system error into the imaging screen, however it was not doing this earlier today and just started doing this. My colleague thinks he hit a button on the underside of the keyboard earlier that reset the system. Since then we have tried to start it up and first got “System Error MO Disk Not Ready.” But now when we boot up the system with the MO Disk in the drive it now shows the system error that “lynx.os.ssp cannot be found.” I was hoping that maybe you guys might have some ideas about what was going on and how we could address this issue to fix it.
Thank you in advance for your help.
Please reply here or email us at andrew@ahrillc.com
I am not sure whether my problem is the same BUT my acuson sequoia do not go into imaging mode....when i switch it on it stacks at 'SYSTEM STARTING.....' but I never starts.
Please Help me
system error that the MO Disk was not ready after just booting the system up while it still has the boot up gray screen with picture of the tree and will not go past this system error into the imaging screen, loading was complete but still shows MO not ready on screen. can you please help?
Hi Marko
I have an Acuson Sequoia C512. When I switch on it, on the screen is displayed the message "System starting ...", but it isn't start ever. All the buttons are "frozen" and to switch off must pull on the plug.
When I put in the new HDD with the same characteristics as the old one, on the screen is displayed massage :ERROR 30".
I can't reload SW because as I mention all the butons are "frozen"
Is SW corrupted and a new HDD must be loaded?
(5)
(1)
Dinu Papoiu
Problem with Acuson Sequoia 512
June 22, 2011 08:57
Hi:
The monitor on the Sequoia will not go on.When I put another monitor on the machine, it starts loading programs but stops and doesn't load anymore.Does anyone have a solution to this problem?
Thank you,
Dinu
to rate and post a comment
(72)
(5)
Marko Shevchuk
Re: Problem with Acuson Sequoia 512
June 23, 2011 06:28
Dinu, hi!
I. You can try: press MULTIHERTZ and V KEY
during power on = you may have more informations
about boot sequence untill loading stops.
II. The most likely causes of a Sequoia system
hang:
1. SSP failure on the RDP board
2. Missing Clock signals from RX. (Secondary
clocks from CN)
3. Faulty Hard Drive
4. Misconfiguration
5 Corrupt data
6. SCSI problem
7. RI or AUX Preamp not seated correctly
I hope this will help you.
Marko
P.S. CAUTION!
I. Do not replace the following
combinations of FRUs at the same time:
1. Hard drive and RDP board
2. Hard drive and PIC board
II. Before replacing an FRU, it is recommended that the system be in
the service enabled mode.
to rate and post a comment
THANH TRUONG
re: Re: Problem with Acuson Sequoia 512
April 15, 2013 10:17
Hi Marko Shevchuk
Could you help me ? My sequoia 512 have applications erro message when booting and it stop boot after that
Please show me how to solve this application erro
My email drthienphuoc@yahoo.com.vn
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: re: Re: Problem with Acuson Sequoia 512
May 06, 2013 04:30
Hi Thanh.
Pls check your mail.
to rate and post a comment
victoriano munoz
re: re: re: Re: Problem with Acuson Sequoia 512
June 09, 2014 04:07
i have the same problem can you help me out?
to rate and post a comment
(35)
(4)
Rey Gonzalez
re: re: re: re: Re: Problem with Acuson Sequoia 512
June 23, 2014 08:22
you need to reload software, contact me rey@allultrasoundservices.com
to rate and post a comment
AHRImedical
re: re: re: re: re: Re: Problem with Acuson Sequoia 512
August 25, 2016 04:36
We are getting the system error that the MO Disk was not ready after just booting the system up while it still has the boot up gray screen with picture of the tree and will not go past this system error into the imaging screen, however it was not doing this earlier today and just started doing this. My colleague thinks he hit a button on the underside of the keyboard earlier that reset the system. Since then we have tried to start it up and first got “System Error MO Disk Not Ready.” But now when we boot up the system with the MO Disk in the drive it now shows the system error that “lynx.os.ssp cannot be found.” I was hoping that maybe you guys might have some ideas about what was going on and how we could address this issue to fix it.
Thank you in advance for your help.
Please reply here or email us at andrew@ahrillc.com
to rate and post a comment
(1)
Simon Mrema
re: re: re: Re: Problem with Acuson Sequoia 512
February 27, 2015 10:17
I am not sure whether my problem is the same BUT my acuson sequoia do not go into imaging mode....when i switch it on it stacks at 'SYSTEM STARTING.....' but I never starts.
Please Help me
to rate and post a comment
(35)
(4)
Rey Gonzalez
re: re: re: re: Re: Problem with Acuson Sequoia 512
March 01, 2015 03:45
contact me rey@allultrasoundservices.com
to rate and post a comment
pedro
re: re: re: re: Re: Problem with Acuson Sequoia 512
April 03, 2016 12:46
Hi Simon How you solve your problem with sequoia before a year ago. I have the same problem now.
Pls give me an answer at agjylameti@gmail.com
to rate and post a comment
FRANK OFORI
re: re: re: Re: Problem with Acuson Sequoia 512
July 15, 2019 10:21
system error that the MO Disk was not ready after just booting the system up while it still has the boot up gray screen with picture of the tree and will not go past this system error into the imaging screen, loading was complete but still shows MO not ready on screen. can you please help?
to rate and post a comment
Kir kavra
re: Re: Problem with Acuson Sequoia 512
June 23, 2015 05:55
Hi , do you have S/W for Acuson Sequoia 512? You coud send me a copy? Manny thanks! My e-mail : muhc08@gmail.com
Donald
to rate and post a comment
pedro
re: Re: Problem with Acuson Sequoia 512
April 07, 2016 03:14
Hi Marko
I have an Acuson Sequoia C512. When I switch on it, on the screen is displayed the message "System starting ...", but it isn't start ever. All the buttons are "frozen" and to switch off must pull on the plug.
When I put in the new HDD with the same characteristics as the old one, on the screen is displayed massage :ERROR 30".
I can't reload SW because as I mention all the butons are "frozen"
Is SW corrupted and a new HDD must be loaded?
Has it problem with PIC or RDP or both?
May you tell me what happened?
to rate and post a comment