Siemens X300 no power up
September 02, 2014 04:35
Hi,
I have problem with Siemens X300. When I press the power on button the machine briefly (like for 1-2 sec) power on and immediately shuts down.
On power supply I can see:
PWR ON - green;
OCP - yellow;
DC - yellow;
lights on.
Does it mean that power supply is bad?
Regards,
to rate and post a comment
Lenny Place
re: Siemens X300 no power up
September 03, 2014 09:01
the LED staus as you describe shows a overvoltage error. can you measure the supplies at the test points
Reference Number Title Test Point Value Tolerance
TP1 P2.5D +2.5V ± 5%
TP2 P5D_PC +5.1V ± 5%
TP3 P5.0D +5.0V ± 5%
TP4 P3.3D +3.3V ± 5%
TP5 M12D -12.0V ± 5%
TP6 P12D +12.0V ± 5%
TP7 P12D_PC +12.0V ± 5%
TP8 PSRTRXVO 0 ~ 18.0V ± 5%
TP9 M100HV -100.0V ± 5%
TP10 P100HV +100.0V ± 5%
TP11 M14AQ -14.2V ± 5%
TP12 P14AQ +14.2V ± 5%
TP13 M2AG -2.0V ± 5%
TP14 P5AQ +5.0V ± 5%
TP15 P3.3AQ +3.3V ± 5%
TP16 GND NA
TP17 P3.3A +3.3V ± 5%
TP20 P5.0V-STBY +5.2V ± 0.5V
to rate and post a comment
(10)
Andrew Weydmann
re: re: Siemens X300 no power up
September 03, 2014 12:54
Thanks for test point reference.
to rate and post a comment
Jefferson Oro
re: re: Siemens X300 no power up
November 19, 2019 04:55
Hello, I have the same problem. But at x150. When you turn it on, it turns off, getting the same lights as the source above. Could someone help me?How can I measure voltages with the power off?
to rate and post a comment
Lenny Place
re: re: re: Siemens X300 no power up
November 20, 2019 09:38
you cannot measure voltage with it off but you can remove boards to see if one of them is loading the supply down. what is you email address i can send you some info
to rate and post a comment
Jefferson Oro
re: re: re: re: Siemens X300 no power up
November 21, 2019 11:15
If I turn off the 48V power, it turns on and stays on, but does not generate the 12V.
to rate and post a comment
Anis Ben slimane
re: re: Siemens X300 no power up
December 23, 2019 10:23
Hi,
Please can you send me the acuson x700 service manual.
Thanks
to rate and post a comment
Majed Salha
re: re: Siemens X300 no power up
June 06, 2022 10:30
HI
i have prroblem with x300 it is not start
the power switch not light
i see in the power supply justthe led pow sw make flacher
to rate and post a comment
(57)
(6)
Jim Carr
re: Siemens X300 no power up
September 03, 2014 09:01
It could be the power supply, or the over current protection may be coming on because some board has failed and is pulling too much current. I would try measuring the resistance of each power supply bus at the test point on the power supply and then removing boards one at a time to see if the resistance changes more than expected for one of them.
TP1 is the +48V output
TP3 is the +12.0V output
We have X300 parts in stock, and people that work on them almost every day that can assist you.
to rate and post a comment
(10)
Andrew Weydmann
re: re: Siemens X300 no power up
September 03, 2014 12:54
Thank you for prompt response and excellent information about pinpointing the issue.
to rate and post a comment
Jefferson Oro
re: re: Siemens X300 no power up
November 29, 2019 01:58
these points Tp1 and tp3, which card are on? Is it in DC or AC / CC?
to rate and post a comment
(35)
(4)
Rey Gonzalez
re: Siemens X300 no power up
September 03, 2014 09:01
It means that the DC power supply is reporting Over current Fault.
rey@allultrasoundservices.com
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: Siemens X300 no power up
September 03, 2014 09:02
Hi Andrzej, yes - DC power has overcurrent range over 150%!!! markoshv@mail.ru I sent you document.
to rate and post a comment
(10)
Andrew Weydmann
re: re: Siemens X300 no power up
September 03, 2014 12:54
Thank you,
I need to do more testing on this machine.
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: re: re: Siemens X300 no power up
September 03, 2014 05:34
I sent you two documents, pls check and answer me.
to rate and post a comment
(1)
Yerzhan Konyspayev
re: re: re: re: Siemens X300 no power up
January 21, 2016 06:03
Hello Marko!
I have the same problem. Can you help me? Please you can email me konyspayev@gmail.com
Please!!! Thank you!
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: re: re: re: re: Siemens X300 no power up
January 22, 2016 08:52
Hello Yerzhan, I sent you mail with doc.
to rate and post a comment
(4)
Vimal Prakash Tripathi
re: re: re: re: re: re: Siemens X300 no power up
May 15, 2019 10:16
Can you please send me a copy of the service manual. Thanks
Vimal
to rate and post a comment
edgar dauzon
re: re: re: re: re: re: Siemens X300 no power up
November 20, 2019 09:38
Hello Mr. Marko,
I have a problem with X300. When power on, the message "The system is shutting down automatically to prevent permanent damage. Please contact Siemens service" pops off.
Can you please guide me through it. Also service manual si highly appreciated.
Thank you.
Edgar
to rate and post a comment
weiming liu
re: re: re: re: Siemens X300 no power up
February 12, 2018 09:17
HELLO Marko,
Could you help send to me for those documents? Thank you! weiming.liu@venture.com.sg
to rate and post a comment
(1)
Yerzhan Konyspayev
re: re: re: Siemens X300 no power up
January 21, 2016 02:46
Hello Andrew!
I have the same problem with the Siemens X300. Can you help me please if you find a problem?
to rate and post a comment
(10)
Andrew Weydmann
re: re: re: re: Siemens X300 no power up
January 22, 2016 01:48
I've ended up replacing the internal ( left side ) dc/dc converter. Unfortunately more expensive part. Be careful about revision there is at least two different ones.
Hope it helps, regards
to rate and post a comment
(1)
Yerzhan Konyspayev
re: re: re: re: re: Siemens X300 no power up
January 24, 2016 04:16
Thank you Andrew Weydmann!
I'll try to find a faulty item.
to rate and post a comment
(3)
Kim Vicmudo
re: re: Siemens X300 no power up
August 16, 2017 10:23
Hello Marko... can you help me?.I am having the same problem now. Thank you! :)
to rate and post a comment
(67)
Gabriel Cool
re: re: re: Siemens X300 no power up
August 17, 2017 10:58
Hi. You need check Dc Dc power suply. +100 V is dosnt work
to rate and post a comment
(72)
(5)
Marko Shevchuk
re: re: re: Siemens X300 no power up
August 17, 2017 04:51
Hello Kim,
pls check your mail
to rate and post a comment
(2)
Keiber Vasquez
re: re: re: re: Siemens X300 no power up
July 27, 2018 09:43
Hi friend? Good afternoon... I have the same problem with the acusom x300. You can send information about equip?? This is my correo keiberarmando2@gmail.com . Thank you very much
to rate and post a comment
Jefferson Oro
re: re: Siemens X300 no power up
November 29, 2019 01:58
What are the points where I should measure this current? and this point is on the DC board?
to rate and post a comment
Jefferson Oro
re: Siemens X300 no power up
November 19, 2019 04:56
Hello, how are you? Got your problem resolved
to rate and post a comment