With Stereoscope II 1998 we have had problems suddenly with picture. First screens were normal black/grey but picture didn't came on the screen. Only memory "received" pictures according the table. After several try we loosed power from screen and there is nothing. What might be problem?
take off the video camera that is over the image intensifier. watch not to direct this unit to a light directly, and focus some thing. You have to see a picture on your screen. If not, the problem is on the video chain. If yes the problem could be on the intensifier tube. With a mirror (in order not to be radiated) do fluoro and you have to see a green circle with image screened. Now, you are able to fix if it is the generator, the intensifier or the video chain.
(1)
Ville Ekroos
Stenoscope II no picture
January 27, 2015 10:24
With Stereoscope II 1998 we have had problems suddenly with picture. First screens were normal black/grey but picture didn't came on the screen. Only memory "received" pictures according the table. After several try we loosed power from screen and there is nothing. What might be problem?
to rate and post a comment
(1)
Ville Ekroos
re: Stenoscope II no picture
January 27, 2015 05:43
Power back on screens but still missing picture
to rate and post a comment
(1)
Ville Ekroos
re: re: Stenoscope II no picture
January 28, 2015 09:55
During and after FLUORO black screen and RAD only same grey before an d after.
to rate and post a comment
(6)
(1)
Raymundo Moreno
re: re: Stenoscope II no picture
January 28, 2015 05:36
take off the video camera that is over the image intensifier. watch not to direct this unit to a light directly, and focus some thing. You have to see a picture on your screen. If not, the problem is on the video chain. If yes the problem could be on the intensifier tube. With a mirror (in order not to be radiated) do fluoro and you have to see a green circle with image screened. Now, you are able to fix if it is the generator, the intensifier or the video chain.
to rate and post a comment
(1)
Ville Ekroos
re: re: re: Stenoscope II no picture
January 29, 2015 10:03
Many many thanks. I will check this. It could be also cable, because socket was a little bit defected.
to rate and post a comment
(1)
Ville Ekroos
re: re: re: Stenoscope II no picture
January 29, 2015 10:04
It looks that we have problem in intensifier or HV power. We don't have any light from tube. do you have spare parts?
to rate and post a comment
(82)
(3)
juan pablo amado
re: re: re: re: Stenoscope II no picture
February 01, 2015 10:26
check your mail
to rate and post a comment
(1)
Ville Ekroos
re: re: re: re: re: Stenoscope II no picture
February 12, 2015 03:59
Problem solved. We managed to fix HV power, only "clock circuit" was broken. Many thanks of all support.