kev

9800 plus c-arm

July 14, 2017 01:29

have a blank left screen, when video cable disconnected says no connection on screen, monitor is on , any advice on why I have nothing on screen, the power supplies check out , have a 7f on video controller showing , shows 5 arrows on boot up on c and stops at 5

Log inor Register

to rate and post a comment

(1)

Raymond Coombs

re: 9800 plus c-arm

July 15, 2017 09:18

check all you fuses on the right (looking from the rear), sounds like you dont have power to the CPU or 4 of the left boards, if you dont have lights on the left half of the rear boards its most likely a fuse

Log inor Register

to rate and post a comment

(24)
(3)

Jesse

re: 9800 plus c-arm

July 15, 2017 09:33

5 Arrows is a loss of connection from monitor cart to generator. Check your interconnect cable. most common is the pins on the cable or the cables inside the connector are broken.