acquisition signal error on Esaote MRI Scan

May 03, 2021 12:56

Hi,
do you have experience about this images?
The MRI is S-scan Esaote
the link to see images is
https://drive.google.com/drive/folders/1VTrNoA9DrMerqoOEzVxhTNbqvH8mCABO?usp=sharing

thanks

Log inor Register

to rate and post a comment

(64)
(14)
(1)

sford

re: acquisition signal error on Esaote MRI Scan

May 05, 2021 10:27

wow, those are weird. I'd start by eliminating things that are NOT the cause. What does the magnet homogeneity look like? Does the problem occur in all orientations? How about when you reverse the phase and readout directions?