AGFA DRYSTAR 300 error MDM sensor

(62)
(10)
(14)

James Fower

AGFA DRYSTAR 300 error MDM sensor

March 20, 2017 05:32

Hi colleagues,
my customer is getting error:
LAST MDM-SENS NOT OK
Any help is appreciated. Service manual, discks
Thank you all in advance

1 person liked this
Log inor Register

to rate and post a comment

(135)
(16)

CRAIG WAMBURA

re: AGFA DRYSTAR 300 error MDM sensor

March 22, 2017 10:24

MDM stands for"MACRODENSITOMETER". It is built-in to allow for constant image quality concerning Dmin, Dmax, sensitometry and vertical banding.

There are no adjustment procedures for the MDM but I would recommend you perform a full calibration of the MDM.

I can help with service manual and service software.
craig@telemed-technologies.com

Log inor Register

to rate and post a comment

(40)
(7)

Ricardo Rivera

re: AGFA DRYSTAR 300 error MDM sensor

March 24, 2017 10:36

hi, the MDM is the densitometer, you can calibrate it in the control panel to level 2 or 3 maybe it everything, if not check the halogen lamp of the mdm, maybe is bad or not clean, it is in the right side of the machine.

Log inor Register

to rate and post a comment

Altima Diagnostic Imaging Solutions - Call (469) 644-7063

Providing dependable, competitive service for Siemens, Philips, and Toshiba MRI and CT. Experts in Oxford Magnet diagnosis and services. Altima offers ramp and shim services for Siemens TIM and many other MRI systems.

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED