AIRIS II CPU not getting ready

Tags: MRI, HITACHI

(5)

Oleksii Riabchenko

AIRIS II CPU not getting ready

August 05, 2014 04:17

Good day. Everything was fine, but today CPU just not getting ready on Airis II. WE have replaced the hdd, but still have the same problem. Please suggest what to do.
nalimca@rambler.ru

Log inor Register

to rate and post a comment

(2)

Ansar Mahdi

re: AIRIS II CPU not getting ready

August 06, 2014 08:09

It is most probably not a CPU problem, rather some hardware module is not reporting ready status to cpu. Gradients? RF? Table? You can check individual component status from main screen. There is an icon to click on top right of screen and from there u can check individual component status.

Log inor Register

to rate and post a comment

(5)

Oleksii Riabchenko

re: re: AIRIS II CPU not getting ready

August 06, 2014 12:10

there are only two components that not ready. they are console and the table. We are usually switch on the gantry only when CPU ready are already not blinking.

Log inor Register

to rate and post a comment

(14)
(5)
(1)

Alex Trayduk

re: AIRIS II CPU not getting ready

August 06, 2014 08:09

Hello,

I had this problem on our MRI last year. I can help you to solved this problem. Write me by my e-mail: trayduk88@mail.ru

Log inor Register

to rate and post a comment

(5)

Oleksii Riabchenko

re: AIRIS II CPU not getting ready

August 07, 2014 10:28

I have found the problem. The broken pcs is an MRPSC board.
is it possible to fix it? or the only way is to replace it?

Log inor Register

to rate and post a comment

(31)
(12)
(2)

Thomas Onyema

re: re: AIRIS II CPU not getting ready

November 08, 2019 10:40

we are having the same problem with our MRPSC Board. Please how did you resolve yours? Did you fix or replace the Board? Thanks a lot.

Log inor Register

to rate and post a comment

(14)

tiseer issa

re: AIRIS II CPU not getting ready

August 07, 2014 10:28

I had resolved this problem last year I thing the problem in the rf amp if the system complete booting