AMX 4 no display

July 21, 2021 09:30

I have an AMX 4 that I changed the battery charge board due to a fried resistor. I took the batteries out and charged each of the 9 batteries to 12.9-13 vdc. Reinstalled the batteries and the unit will not show anything on the display. The K 216 relay clicks but nothing else.

Log inor Register

to rate and post a comment

(1)

michael CARABOTTA

re: AMX 4 no display

July 22, 2021 10:52

Check and see if it makes an exposure. If it does replace the kv/ma display panel. (the small one that you increase/decrease kv/ma on). When this board goes bad it shorts out both panel displays. The long display board will still be ok.

Log inor Register

to rate and post a comment

(21)
(1)

Charley Russell

re: re: AMX 4 no display

July 23, 2021 11:52

Tried that, no rotor or xray. Tried two different controller boards, no go. Checked the power supply, and all of the outputs are good. Lights on controller board come on. The unit will not drive either.

Log inor Register

to rate and post a comment

X-Ray Tube Replacements for GE, Siemens, Varian, Philips, Toshiba, Shimadzu

Over thirty five years in the X-Ray tube manufacturing. X-Ray tube replacements for all leading imaging equipment manufactures. From the past twenty years to today's modern mobile digital imaging equipment. Overnight shipping. SC, USA. 843 767 8090

advertisement

 

Hitachi

 

RSTI Newsletter

Upcoming Auction Events

ZRG Medical
TIMED Ending Soon
Aug 06, 2021 (77 lots)
Tekyard Medical
TIMED
Aug 11, 2021 (547 lots)
MedSurge LLC
TIMED
Aug 12, 2021 (85 lots)
KEEBOMED
TIMED
Aug 16, 2021 (400 lots)
Alternup Medical
TIMED
Aug 24, 2021 (163 lots)