December 06, 2017 08:52

On Toshiba Aquilion 16, When performing a Helical scan I can see a ring artifact near approx an inch to outer circumference in the TOS phantom. Performed NRA, air calibrations and scano cal but still the issue persists. Please advice. My email id is sri.seelam@rmedservices.com

Thanks for everyone in advance.

Log inor Register

to rate and post a comment

(28)
(8)

Tolga GULUMSER

re: Artifact in Helical scan in a Toshiba Aquilion 16

December 20, 2017 09:16

It can be defective QV2 or ADC2 boards. If it is single ring you can try to delete channel.