> BrightSpeed - Scan was Paused/Aborted by Host

BrightSpeed - Scan was Paused/Aborted by Host

Tags: CT, GE

(116)
(27)

FranciscoG

BrightSpeed - Scan was Paused/Aborted by Host

March 17, 2017 05:22

Hi guys,

I'm working on this BrightSpeed that was working and suddenly it aborts the scans, I'm able to warm up the tube, and the only messages that show in the error log are the following:

Host: tac1 Proc: scanmgr Error:245213
File: scanmgr_events.cxx
Function: System Control: Scan Control
Scan Type: Axial Scan:50000/1/3
Scan Seq Id: 1100
Scan was Paused/Aborted by Host

Host: tac1 Proc: scanmgr Error:245190
File: scan_complete.cxx
Function: System Control: Scan Control
Scan Type:Axial Scan 0/0/0
Scan Seq Id: 1100
System Aborted Axial scanning
No views, either offset or data, were collected/recieved.
Data views expected:992


So I can move table, positioning the craddle and as soon as I hit START the scan, the gantry rotates and you could hear the rotor running as well, but the system aborts the scan before it even makes the xray exposure.

I found that the ORP2 board, has this 2 leds lighted in RED:

ERROR: RED LED
ERR HIST: RED LED

When I try to reset the ORP2 board, once is reseted, this 2 RED leds come on again. I checked the emergency switches and they are good, as well as the tape switches.

I do not know how to test the ORP2 board if there is a way on the system, please if someone has any insight on this let me know.

Thanks.

3 people liked this
Log inor Register

to rate and post a comment

(27)
(7)

HECTOR VELAZQUEZ

re: BrightSpeed - Scan was Paused/Aborted by Host

March 19, 2017 04:59

Hello francisco. The information you uploaded is not enough to know what's going
On your machine. Send me an email and i will help you hvelazquezm@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(27)
(8)
(5)

Vijay Sharma

re: BrightSpeed - Scan was Paused/Aborted by Host

March 19, 2017 04:59

Few more errors logs expected here in this case. pls share the error log file or error log screen shot to look further.

1 person liked this
Log inor Register

to rate and post a comment

(13)
(3)

Ignacio Garcia

re: BrightSpeed - Scan was Paused/Aborted by Host

March 19, 2017 05:00

You have a problem with the data views, check your slip rings, your fiber optic.
Run Scan data path to see if you have a good data reception...

1 person liked this
Log inor Register

to rate and post a comment

(1)
(1)

Paul66

re: BrightSpeed - Scan was Paused/Aborted by Host

March 19, 2017 05:02

Try to do the Flash download to see where is the problem.
if not help try to do LFC on Host PC

Pawel

1 person liked this
Log inor Register

to rate and post a comment

(116)
(27)

FranciscoG

re: re: BrightSpeed - Scan was Paused/Aborted by Host

March 27, 2017 10:03

Hi Paul66, the Flash download was succesfully updated, and even the LFC was performed as well, and keeps doing the same thing.

1 person liked this
Log inor Register

to rate and post a comment

(44)
(8)

Pawel Kopysc

re: re: re: BrightSpeed - Scan was Paused/Aborted by Host

March 28, 2017 09:23

Hi,

Have You checked the logs? There must be an error number. If You do the flash download the ORP is working properly?
Try to do some X-ray Gen. test to see if You can do scan parameters like kV mA etc.

You can also run Scan data path, check slipring communication

As i said there should be an error in log viewer

Pawel

1 person liked this
Log inor Register

to rate and post a comment

(84)
(7)

CRAIG WAMBURA

re: BrightSpeed - Scan was Paused/Aborted by Host

March 20, 2017 10:32

Do you have any "Taxi Modem" errors?
craig@telemed-technologies.com

1 person liked this
Log inor Register

to rate and post a comment

(116)
(27)

FranciscoG

re: re: BrightSpeed - Scan was Paused/Aborted by Host

March 30, 2017 08:27

Hi Craig, no I have no taxi modem errors, thanks man.

Log inor Register

to rate and post a comment

(116)
(27)

FranciscoG

re: BrightSpeed - Scan was Paused/Aborted by Host

March 30, 2017 08:27

Thanks to everyone, I'm stil working on this system, these are so far the tests and results I've performed on it.

0. FLASH Download Tool

Query Update

CCB OK CCB OK
DCB OK DCB OK
JEDI OK JEDI OK
ORP OK ORP OK
TCB OK TCB OK
TGP OK TGP OK

1. Scan Data Path

TEST 1 TEST 2 TEST 3 TEST 4
Data Path Selection: DCB DCB Conv. Conv.
Trigger Source: Axial Bd DCB Axial Bd DCB
X ray: OFF OFF ON ON
Iterations: 3 3 3 3
Gantry: Disabled Disabled Disabled Disabl
Completed YES YES YES YES

Scan Aborted YES YES YES YES


TEST 5 TEST 6 TEST 7 TEST 8
Data Path Selection: DCB DCB Convertor Convertor
Trigger Source: Axial Bd DCB Axial Bd DCB
X ray: OFF OFF ON ON
Iterations: 3 3 3 3
Gantry: Enabled Enabled Enabled Enabled
Completed YES YES YES YES

Scan Aborted YES YES YES YES


2. DAS Tools
Message:
Auto Test Scan Hardware Stopped Scan

3. DDC
Message:
X ray OFF (Static) Scan Hardware Stopped Scan
X ray ON (Static) Scan Hardware Stopped Scan
X ray OFF (Rotating) Scan Hardware Stopped Scan
X ray ON (Rotating) Scan Hardware Stopped Scan

4. I've changed the IG and the DARC nodes just because I have them from a good working system, but the same errors are still present. And the DS5 (ERROR) and DS6 (HIST ERR) leds from the ORP2 are lighted in orange.

****
These are the only errors appearing in the service log:

Host: tac1 Proc: scanmgr Error:245213
File: scanmgr_events.cxx
Function: System Control: Scan Control
Scan Type: Axial Scan:50000/1/3
Scan Seq Id: 1100
Scan was Paused/Aborted by Host


Host: tac1 Proc: scanmgr Error:245190
File: scan_complete.cxx
Function: System Control: Scan Control
Scan Type:Axial Scan 0/0/0
Scan Seq Id: 1100
System Aborted Axial scanning
No views, either offset or data, were collected/recieved.
Data views expected:992


Host: tac1 Proc: scanmgr Error:245177
File: scanmgr_events.cxx
Function: System Control: Scan Control
Scan Type:Scout Scan 0/0/0
Scan Seq Id: 1100
System Aborted Scout scanning
An Acquition Complete event was received when Scan Manager was not in a scanning state.
Acq Complete Sequence number: 7000

Thanks again to everyone for your insights, they are helpful.

Log inor Register

to rate and post a comment

Marquis Medical - Experts in SIEMENS & GE PET & PET/CT Service & Parts

Our engineers are trained strictly on Siemens & GE PET to Guarantee an expert on all service calls. Our large inventory of PET & PET/CT systems & parts allows us to provide a TRUE Turn-key solution. Ph:866-645-7457 DOTmed Certified

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED