doug westlake

bucky diagnost ts table

November 08, 2017 09:27

i have a bucky diacnost ts table stuck in the highest position,i see the relays pull in on the controller pcb..but the motor still does not drive..i;m told it is the encoder in motor..all swiches ohmed out and only 1 green led on board. can i manually move the motor?i dont see how.

Log inor Register

to rate and post a comment

medo alhadad

re: bucky diagnost ts table

November 09, 2017 08:31

chek calibration sum error pleas help What's wrong Do you have software? chemistry analyzer biosystem 310

Log inor Register

to rate and post a comment

(57)
(5)

Juan Antonio Ramirez Ramirez

re: bucky diagnost ts table

November 09, 2017 08:36

Did you tried to pull out table tray? I mean if you pull out table tray this hooks an actuator that prepares table bucky for next exposure

Log inor Register

to rate and post a comment

(7)

(13)

Gustavo Polo

re: bucky diagnost ts table

November 09, 2017 01:43

we have parts for this table if you need.

Log inor Register

to rate and post a comment

(294)
(27)

DMEDS

re: bucky diagnost ts table

November 13, 2017 05:41

i have manuals for this table if you need it contact me,
magmesme@gmail.com
regards

Log inor Register

to rate and post a comment

Technical Assistance & Parts for CT Scanners 24h/7 days Call 813-925-1314

ElsMed engineers are located in four different countries. 24 hour response time to any service call. Service agreement provided for end-users range from a 6 month period to a 3 year period and covers all system spare parts. DOTmed Certified DOTmed 100

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED