doug westlake

bucky diagnost ts table

November 08, 2017 09:27

i have a bucky diacnost ts table stuck in the highest position,i see the relays pull in on the controller pcb..but the motor still does not drive..i;m told it is the encoder in motor..all swiches ohmed out and only 1 green led on board. can i manually move the motor?i dont see how.

Log inor Register

to rate and post a comment

medo alhadad

re: bucky diagnost ts table

November 09, 2017 08:31

chek calibration sum error pleas help What's wrong Do you have software? chemistry analyzer biosystem 310

Log inor Register

to rate and post a comment

(72)
(5)

Juan Antonio Ramirez Ramirez

re: bucky diagnost ts table

November 09, 2017 08:36

Did you tried to pull out table tray? I mean if you pull out table tray this hooks an actuator that prepares table bucky for next exposure

Log inor Register

to rate and post a comment

This Comment has been deleted

November 09, 2017 01:43

Log inor Register

to rate and post a comment

(347)
(30)

DMEDS

re: bucky diagnost ts table

November 13, 2017 05:41

i have manuals for this table if you need it contact me,
magmesme@gmail.com
regards

Log inor Register

to rate and post a comment

(30)
(4)
(1)

doug westlake

re: re: bucky diagnost ts table

December 10, 2017 12:06

i would appreciate any info i can get.

Log inor Register

to rate and post a comment

RaySafe helps you avoid unnecessary radiation

RaySafe solutions are designed to minimize the need for user interaction, bringing unprecedented simplicity & usability to the X-ray room. We're committed to establishing a radiation safety culture wherever technicians & medical staff encounter radiation.

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED