BV Endura Error M351 and M208

odilon cardoso

BV Endura Error M351 and M208

February 11, 2016 11:47

BV Endura Error M351 and M208, Can you help me?

1 person liked this
Log inor Register

to rate and post a comment

(2683)
(629)
(19)

Lenny Place

re: BV Endura Error M351 and M208

February 11, 2016 05:53

M 351 FA_INVERTER_A LARM X-ray generator not available
Switch system off/on Call service if problem is persistent FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.


M 208 XACT_GONE Subsystem error,
system still available It's possible to continue Unexpected X-ray switch off. Check board SHA4 and the generator unit SE. Check cabling.

Log inor Register

to rate and post a comment

Dean Smith

re: re: BV Endura Error M351 and M208

February 12, 2016 09:01

U have a bad HV tank!! Replace it... Test: fails in High KV!!! Doesnt fail in low KV!!! Replace and calibrate system! U will beed BV scope and philisp Ist!!!

Regards,

Dean

Log inor Register

to rate and post a comment

odilon cardoso

re: re: re: BV Endura Error M351 and M208

February 12, 2016 01:51

When put in or more 65kV, ocurre this error, but in 40 45 50 and 60kv no problem.

Log inor Register

to rate and post a comment

(2683)
(629)
(19)

Lenny Place

re: re: re: re: BV Endura Error M351 and M208

February 12, 2016 03:40

bad mono block

Log inor Register

to rate and post a comment

(2)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How to fix errors M351 and M208

Log inor Register

to rate and post a comment

(2)

Kirill Severtsov

re: re: re: re: re: BV Endura Error M351 and M208

June 07, 2018 09:42

How did you fix the errors M208 and M351 of BV Endura 2.3

Log inor Register

to rate and post a comment

Basheer Almetwakel

re: re: re: BV Endura Error M351 and M208

June 03, 2021 10:28

How to calibration it please help me to calibration it thanks

Log inor Register

to rate and post a comment

DMEDS

re: BV Endura Error M351 and M208

February 12, 2016 09:01

351
-
x ray generator not available Switch system off/on.
FA generator board SHA4 Inverter status signal is active. Generator is reset within a few seconds. Check the kV/mA values of the tank.

208

Subsystem error, system still available. It's possible to continue
Unexpected X ray switch off. Check board SHA4 and the
generator unit SE. Check cabling.

if you need complete service manuals contact me,

magmesme@gmail.com

Regards

1 person liked this
Log inor Register

to rate and post a comment

lin

re: re: BV Endura Error M351 and M208

June 12, 2018 08:38

Mr i need the service manuals ,pls help me,send the manuals to the srey123@163.com,thanks

Log inor Register

to rate and post a comment

(2683)
(629)
(19)

Lenny Place

re: re: re: BV Endura Error M351 and M208

June 12, 2018 11:09

I am sending using myairbridge

Log inor Register

to rate and post a comment

lin

re: re: re: re: BV Endura Error M351 and M208

June 13, 2018 07:44

Can you send the manual by srey123@163.com ,thanks

Log inor Register

to rate and post a comment

(2683)
(629)
(19)

Lenny Place

re: re: re: re: re: BV Endura Error M351 and M208

June 14, 2018 01:44

Sending again

2 people liked this
Log inor Register

to rate and post a comment

lin

re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 07:59

Thank you !I have received the manual,the manual in the junk mail.

Log inor Register

to rate and post a comment

(2683)
(629)
(19)

Lenny Place

re: re: re: re: re: re: re: BV Endura Error M351 and M208

June 15, 2018 05:13

Hope it helps you are welcome

1 person liked this
Log inor Register

to rate and post a comment

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com Call 813-925-1314

DOTmed Certified DOTmed 100
advertisement

 

Hitachi

 

RSTI Newsletter

Upcoming Auction Events

Tekyard Medical
TIMED
Jun 16, 2021 (289 lots)
Primis Medical
TIMED
Jun 18, 2021 (122 lots)
Alternup Medical
TIMED
Jun 21, 2021 (143 lots)
Medical Equipment Superstore
TIMED
Jun 22, 2021 (570 lots)
Primis Medical
TIMED
Jun 24, 2021 (221 lots)
ZRG LLC.
TIMED
Jun 25, 2021 (101 lots)