Bv Endura Error M393 - Service Technicians Forum

Over 1300 Total Lots Up For Auction at Four Locations - NJ Cleansweep 03/02, OK 03/04, NJ 03/08, CA 03/12

BV ENDURA ERROR M393

Tags: C-Arm, PHILIPS

(21)
(1)

Celio Dibax

BV ENDURA ERROR M393

February 19, 2021 10:41

I HAVE ERROR M393 IN C-ARM BV ENDURA PHILIPS
(DOSELINK ERROR, X-RAY ABORTED)
WHEN X-RAY IN AUTOMATIC MODE

THANKS IN ADVANCE

CELIO

Log inor Register

to rate and post a comment

(5)

EDUARDO BUSTOS

re: BV ENDURA ERROR M393

February 22, 2021 10:42

Hi,
Error 393 is usually due to the cable.

Log inor Register

to rate and post a comment

(53)
(7)

Willi

re: BV ENDURA ERROR M393

February 22, 2021 10:42

Hi Celio,

This is a problem by the stand trolley cable and/or wire braket by the stand.
This signal comes from the DFI3 via standtrolley cable to the stand and make the regulation of the dose.
Mostly you can move the cable and then you see the error or not.
Kind regards
Willi

1 person liked this
Log inor Register

to rate and post a comment

(5)

EDUARDO BUSTOS

re: re: BV ENDURA ERROR M393

February 22, 2021 11:07

Symptom : Reliability problems regarding the Stand-MVS connection of the BV Family R2
Systems. This can cause pixel faults, error M393/IDSLINK errors or spontaneous
reboot of BV Family R2 Systems. Most of the time the spontaneous reboot is caused
by a SW fault. Therefore FCO71800022 (SW Rel. 2.2.4) has been issued.
Cause : Connection and/or cabling problems.
Remedy : Solve the connections problems by replacing the Stand-MVS cable by an improved

Log inor Register

to rate and post a comment

(53)
(7)

Willi

re: re: re: BV ENDURA ERROR M393

February 23, 2021 10:09

Dear Eduardo,

Yes, this with the FCO is correct, but a very old story and also there was more FCO's by Philips for this issue.

You write the software 2.2.4 will solved some bugs, correct, but 2.2.8 are the last version for 2.2 Systems.

But we don't know from Celio what for a system he has. It can be a 2.1 / 2.2 / 2.3 Hardware System and by all this system the error is possible.

But Philips has changes some importend thinks and the Error is not anymore visible in the field.

Kind regards
Willi

Log inor Register

to rate and post a comment

(5)

rockbob

re: BV ENDURA ERROR M393

February 23, 2021 10:11

check cable between DFI and the SHA1 board in the SUCO

Log inor Register

to rate and post a comment

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions March 8th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 178 lots of medical equipment. Onsite or virtual inspection available. Contact Scott at +1 (856) 655-8055 or email ssfitz34@gmail.com for an appointment.

advertisement

 

RSTI Newsletter

 

Hitachi

Upcoming Auction Events

Premier Medical Depot
TIMED
Mar 04, 2021 (689 lots)
Therapeutic Technologies
TIMED
Mar 08, 2021 (178 lots)
ZRG LLC.
TIMED
Mar 12, 2021 (146 lots)