> BV Libra error m050

BV Libra error m050

Tags: C-Arm

(146)
(10)

Luis Fernando Garcia Sanchez

BV Libra error m050

August 08, 2017 10:53

At start up show in display error M050,I am looking BV scope service program, can you help me?, email Luisfernandogarcias@yahoo.com regards
Luis F. Garcia

Log inor Register

to rate and post a comment

(9)

Willi

re: BV Libra error m050

August 10, 2017 04:33

Dear Luis,

The Error 050 ist a BIST Error.

This error number says nothing because it is displayed as the last error message.

But you can see with the "I" button at the stand more and then the list is displayed. You can then also scroll the list with two buttons at the stand monitor.

This is only possible if the error is up-to-date.

BV-Scope is of no use to you, is protected by the IST system.

If you send me the error numbers, I look gladly.

Many greetings
Willi