Bv Pulsera C-Arm - Service Technicians Forum

bv pulsera c-arm

(4)
(1)

kevin mattingly

bv pulsera c-arm

October 06, 2020 10:16

bv pulsera giving errors m208 and m343. Info says problem with sha4 pcb or cabling. is cabling the interconnect cable or another cable.

Log inor Register

to rate and post a comment

(49)
(6)

Willi

re: bv pulsera c-arm

October 07, 2020 11:13

Hi,

Have you this Errors by high KV? Maybe by lateral Position of the C-Arm?

KR Willi

Log inor Register

to rate and post a comment

(4)

rockbob

re: bv pulsera c-arm

October 12, 2020 09:50

check for fan spinning and for dust or debris blocking opto that monitors fan spinning

Log inor Register

to rate and post a comment

(63)
(17)

tim cowell

re: bv pulsera c-arm

October 14, 2020 10:03

If you cannot get a successful performance test And you are getting these errors. The most likely cause is gonna be the x-ray tube.

Log inor Register

to rate and post a comment