Alberico Velasquez

C-arm Siremobil Compact L

October 12, 2017 09:54

I have a Siremobil Compact L
When the equipment is turned on the display shows these errors
7150 and 6100
And when fluro is made generally above 70 KV some of these pairs of errors appear 4028-5901 or 4028-5030
Can someone tell me the meaning of these errors.
And what can I check on the equipment?
You can send me the information to my email albetoma2611@yahoo.es
Thanks

Log inor Register

to rate and post a comment

(438)
(54)
(1)

Mohammed Abdullah

re: C-arm Siremobil Compact L

October 12, 2017 01:49

Dear Sir,
I have all manuals for this system ,
if you need it contact me,
magmesme@gmail.com
Regards

Log inor Register

to rate and post a comment

(438)
(54)
(1)

Mohammed Abdullah

re: C-arm Siremobil Compact L

October 12, 2017 01:49

Error meaning as follow,
7150
​E7150 Slot diaphragm rotation index not found
Error E7150 Slot diaphragm rotation index not found
Possible trouble: The slot diaphragm rotation index position was not found
during initialization.
Possible help: Check the index switch
Check the index switch wires.
Check the motor.
Check the motor wires.
Check the mechanical movement.
Board D1 defective.
​
6100
​E6100 Generator needs learning
​Error E6100 Generator needs learning
Possible trouble: The following generator learning datas are invalid: .
kV offset.
Push factor,
Filament heating factor.
Possible help: Start Generator learning.

4028
E4028 Memoskop unexpected acq end signal
Error E4028 Memoskop unexpected acq end signal
Possible trouble: Host registered unexpected acq end signal of the Memoskop.
Possible help: Check faults on the Memoskop acquisition cable (D1.X10.27).
Check Memoskop.
Check connection cables to the Memoskop.

5901
Error F5901 Gen. tried X-Ray on without XRAY_query signal
Possible trouble: Generator attempts radiation but the signal
XRAY_query was not present.
Software error D1.
Possible help: Faults in the system.
Sirephos defective
D1 defective.

5030

F5030 Gen. tube current actual value too high
Error F5030 Gen. tube current actual value too high
Possible trouble: Actual tube current (Iroe_ist) too high.
Possible help: Sirephos tube arcing, Sirephos no longer high-voltage-proof.
Measure Iroe_ist at D1."mA".
Check actual value cables.
Measured value acquisition on D1 defective.
Sirephos defective.

wait your feedback

With Best Regards
Good Luck

Log inor Register

to rate and post a comment

(210)
(27)
(27)

Wayne Horsman

re: C-arm Siremobil Compact L

October 13, 2017 09:29

7150 is generally a stripped ring gear on the slot diaphragm. We have replacements, on an exchange basis, if you need one.
4028 is a Memoskop error. We have replacement Memoskops, on an exchange basis, if you need one, but (see below) maybe your Memoskop is OK.
5091 is either a bad x-ray tube (Sirephos) or a D-1 PCB (cpu) We have replacement D-1 PCBs, on an outright basis, if you need one, but (see below) maybe your D-1 is OK.
5030 is usually an arcing x-ray tube (Sirephos). If the 4028 comes after the 5030, 5091, then the Memoskop is probably OK, just a bad Sirephos.

Log inor Register

to rate and post a comment

The Intuitive Biomedical Solution - GE OEC C-Arm Sales, Service & Parts

INTUITIVE THINKING • INTUITIVE WORKS • INTUITIVE SOLUTIONS • Contact TIBS today for your GE OEC C-arm Sales, Service or Parts. Parts & systems in stock now. Exchange or sell us your old C-arm. Call 866.499.3966 or email sales@tibscorp.com.

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED