Nuruddeen Mohammed Abdulbaqi

"CANNOT CONNECT TO RTM - STARTUP FAILED"

July 14, 2017 01:29

I just installed a refurbished Toshiba Aquilion 16. The challenge am facing is that the CPU System Complaints of "STARTUP FAILED - CANNOT CONNECT TO RTM", also the Patient Couch don't respond, as if it does not receive power.
Kindly assist me to find my way out of these problems.

1 person liked this
Log inor Register

to rate and post a comment

(155)
(22)

CRAIG WAMBURA

re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 15, 2017 09:33

Check the LED status of MHR boards of the REC-BOX and see if the LEDs of an MHR-FC is off. If off, Replaced the MHR-FC board in Rack #1


1 person liked this
Log inor Register

to rate and post a comment

(24)
(11)

Nuruddeen Mohammed Abdulbaqi

re: re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 17, 2017 10:18

Thank You Graig. I'll do just that and observe.
Thanks please.

Log inor Register

to rate and post a comment

(155)
(22)

CRAIG WAMBURA

re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 15, 2017 09:33

Check the LED status of MHR boards of the REC-BOX and see if the LEDs of an MHR-FC is off. If off, Replaced the MHR-FC board in Rack #1

1 person liked this
Log inor Register

to rate and post a comment

(67)
(9)

Juancho

re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 16, 2017 03:41

Hi, Nuruddeen
You have service Manual? Check LEds Status on RTM board. You can connected it via RS-232 or RJ-45 with windows Hyper-terminal?

1 person liked this
Log inor Register

to rate and post a comment

(24)
(11)

Nuruddeen Mohammed Abdulbaqi

re: re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 17, 2017 10:18

Thanks Juancho.
I'll do as you advised and make observations.
Thanks alot.

Log inor Register

to rate and post a comment

(24)
(11)

Nuruddeen Mohammed Abdulbaqi

re: CANNOT CONNECT TO RTM - STARTUP FAILED

July 18, 2017 07:46

Thanks for your support.
I've checked the MHR boards, it appeared the 5Amps fuses on boards of slot 10 and 12 are broken broken. Also the "FAULT" LED-0 lits.
Thanks.

Log inor Register

to rate and post a comment

ISS - Contact Us Today For All Your Contrast Injector Needs: 888-667-1062

Injector Support & Service (ISS) offers telephone technical support, depot repair services, loaner injectors, parts identification and sales, PM tools and PM parts kits for medical contrast injectors.

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED