Checkup failure Siemens Somatom Scope 2017

Tags: CT, SIEMENS

Marius Stoica

Checkup failure Siemens Somatom Scope 2017

November 16, 2021 01:56

Hello
I just encountered a problem when performing checkup procedure to a Somatom Scope Power from 2017. During the procedure is entering in a loop (repeating last scan until fatal error ). Any idea? Thank you in advance!

1 person liked this
Log inor Register

to rate and post a comment

(73)
(8)
(7)

JIFFY PAUL AKKARAKARAN

re: Checkup failure Siemens Somatom Scope 2017

November 17, 2021 10:11

Please check the event log. You can get details of the error. If it is aborting scan in warm up cycle then it could be either x-ray tube or the x-ray generator issues. Is aborting in filament adaptation cycle or defective channel correction or in the air calibration cycle?

Log inor Register

to rate and post a comment

(322)
(63)

Gabriel Cool

re: Checkup failure Siemens Somatom Scope 2017

November 22, 2021 11:26

When you have got this error? When system start filament adaptation or deffective channels? If first - then problem can be with tube. If second- check detector or irs

Log inor Register

to rate and post a comment

ISS - Contact Us Today For All Your Contrast Injector Needs: 888-667-1062

Your comprehensive contrast injector solution for 10 years to include: on-site service, telephone technical support, depot repair with loaner options, parts sales, injector sales, injector install & de-install, Level 2 & beyond training.

advertisement

 

RSTI Newsletter

Upcoming Auction Events

KEEBOMED
TIMED
Nov 30, 2021 (1646 lots)
Tekyard Medical
TIMED
Dec 01, 2021 (361 lots)
ZRG Medical
TIMED
Dec 03, 2021 (122 lots)
Primis Medical
TIMED
Dec 09, 2021 (564 lots)