ERROR 02HV & 02HX ON PHILIPS BUCKY

Tags: Radiology - R/F, PHILIPS

(2)
(1)
(1)

John Kisinga

ERROR 02HV & 02HX ON PHILIPS BUCKY

November 22, 2021 11:32

having Errors 02HV when I try to Expose

once I press Reset, system exposes and has Image though I still see new error 02HX

the system is aphilips Bucky Diagnost TH.

anyone who is experienced on this

we Just replaced the HV tank but no Luck

Log inor Register

to rate and post a comment

(269)
(40)

CRAIG WAMBURA

re: ERROR 02HV & 02HX ON PHILIPS BUCKY

November 23, 2021 05:02

This indicates that the divider test anode is out of range

1. check power supply voltage EZ130 (+15, -15, +26)
2. If measurements do not tally, replace PCB

Log inor Register

to rate and post a comment