Error 1579 In Omcp Integris Philips Generator - Service Technicians Forum
advertisement
Current Location:
>
>
>
> Error 1579 in OMCP (Integris) Philips Generator

Error 1579 in OMCP (Integris) Philips Generator

Tags: Cardiac Imaging

(19)
(1)

JUAN ESPINOZA

Error 1579 in OMCP (Integris) Philips Generator

May 30, 2019 09:07

Hello, System is Ok at standby (no error) however at fluo or cine NO images and FNMD says "Action" for a while, then recovery Standby "OK". Error trace shows E1579 ("EXON is missing"). EQ Converter seems inactive (no HF noise), Filament boards are OK. MRC is OK. RDY1 and RDY2 leds are lit (normal) in EZ112 during Xray.
I have replaced EH116, EH102, EZ112 w/o changes.
Any further suggestions?

1 person liked this
Log inor Register

to rate and post a comment

(5)

Camanducaia legueto

re: Error 1579 in OMCP (Integris) Philips Generator

July 19, 2019 10:19

See 1576 error; Possibly defective EVK3 relay in powerpart;
Jumper on EZ112 should be in position 2-3 in case of MRC gridswitch!;

Log inor Register

to rate and post a comment