Need explanation of error 3249 on Philips SUPER 50/CP. Reset from console can not clear error.
Thank You in advance.

Log inor Register

to rate and post a comment

(259)
(50)
(3)

Jose Vallejos

Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

August 20, 2009 03:31

Error 3249

kV set point > ±4% tolerance from EZ133 X22, adjustment R2 1 V - 20 kV measured after every desk data change

Perform:

• Check that EZ137 S1 is in NORM position.
• Check zero adjustment EZ137 R2 (measure at R11) R1 ADZA.
• Check input scanner on EZ137.
• Program error code to ERBUS (error buffer system, service
mode 7.4 and 7.3) and check if this error will be followed by another one (this error code + 1). If so, replace the board.
• Check that filament current limits on EZ126 are properly adjusted.

If need service manual contact me :
javv2005@hotmail.com

Log inor Register

to rate and post a comment

(47)
(4)

Milos Krstic

Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

August 26, 2009 11:13

Thank You for help. I solved problem by adjustment R2 on EZ133.

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 12, 2018 03:19

please , help me for explanation error 3264 at Philips super 50 CP OR send me the service manual :)

Log inor Register

to rate and post a comment

(321)
(30)

DMEDS

re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 10:35

Dear Amr,
i have the manuals for super cp generator ,
if you need it contact me for making offer for you,
magmesme@gmail.com

error 3264 :
Standby filament current, actual value small focus
from EH141 X1B:02 > ±35% tolerance measured
every second in standby and fluoro, during self test
after switch on

thanks

with best regards

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 04:22

I contact with you sir to receive service manual from my mail
bme.amrelsayed@yahoo.com

Log inor Register

to rate and post a comment

(8)

Amr Elsayed

re: re: re: Re: Error 3249 on X-ray PHILIPS SUPER 50/CP

January 17, 2018 04:22

Dear Dmeds,

thank you for offer you help to me , I Contact with you at your email from my mail "" bme.amrelsayed@yahoo.com "" please be confirm send the service manual to me .
thank you again

Log inor Register

to rate and post a comment

Trusted Siemens MRI CT PET/CT & Mobile Parts & Service Specialists

* Premier Siemens equipment & service & part provider * Established in 1995 * World wide reputation * The same day parts delivery* Please call (828) 333-7700! DOTmed Certified DOTmed 100

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED