Amr Elsayed

Error 3264 at philips super 50 CP

January 12, 2018 03:19

I am Facing a problm with philips super 50/cp with error 3264 and i don`t have the service manual , can any one help for solving ,please

contact me at
bme.amrelsayed@yahoo.com

Log inor Register

to rate and post a comment

(379)
(33)

DMEDS

re: Error 3264 at philips super 50 CP

January 17, 2018 10:35

Dear Amr,
i have the manuals for super cp generator ,
if you need it contact me for making offer for you,
magmesme@gmail.com

error 3264 :
Standby filament current, actual value small focus
from EH141 X1B:02 > ±35% tolerance measured
every second in standby and fluoro, during self test
after switch on

thanks

with best regards