Error 35 Mobilett Xp Hybrid - Service Technicians Forum

error 35. Mobilett XP Hybrid

Tags: Radiology - R/F, AMERICAN X-RAY

(3)
(1)

osmarys gonzalez

error 35. Mobilett XP Hybrid

February 21, 2020 10:29

hi, I have a problem with Siemens Mobilett XP Hybrid, error 35 when making an exhibition. Can someone help me with the service manual? Thank you.

1 person liked this
Log inor Register

to rate and post a comment

(5)
(3)

GABRIEL GUTIERREZ

re: error 35. Mobilett XP Hybrid

February 24, 2020 09:31

Hi Osmarys.
This is the error information:
ERR 35 Short-circuit at the main inverter
System mode:

Exposure Explanation: D916 (CPU) detects a main inverter short-circuit.
Action:
· Check test point TP - SCM on D916 (CPU) after exposure, (high = error).
· ERR 35 with exposures in all kV ranges indicates a probable single tank defect.
· Check the cable connections between D916 and D962 (X20).
· Check the cable connections between D962 and the single tank (U - V).
· Check C1.

1 person liked this
Log inor Register

to rate and post a comment

(101)
(20)

Germni

re: error 35. Mobilett XP Hybrid

February 24, 2020 09:32

hello,
ERR35
Short-circuit at the main inverter
Exposure
Explanation: D916 (CPU) detects a main inverter short-circuit.
Action:
• Check test point TP - SCM on D916 (CPU) after exposure, (high = error).
• ERR 35 with exposures in all kV ranges indicates a probable single tank
defect.
• Check the cable connections between D916 and D962 (X20).
• Check the cable connections between D962 and the single tank (U - V).
• Check C1.

1 person liked this
Log inor Register

to rate and post a comment

(101)
(20)

Germni

re: error 35. Mobilett XP Hybrid

February 24, 2020 09:32

Short-circuit at the main inverter

Explanation: D916 (CPU) detects a main inverter short-circuit.
Action:
• Check test point TP - SCM on D916 (CPU) after exposure, (high = error).
• ERR 35 with exposures in all kV ranges indicates a probable single tank
defect.
• Check the cable connections between D916 and D962 (X20).
• Check the cable connections between D962 and the single tank (U - V).
• Check C1.

1 person liked this