> Error 47 (E-47) in the Remstar /pro2 humidifier

Error 47 (E-47) in the Remstar /pro2 humidifier

JAIME ALBERTO GONZALEZ ZULUAGA

Error 47 (E-47) in the Remstar /pro2 humidifier

December 20, 2007 01:29

I have a humidifier Remstar /pro2 S/module 1005792 S/N H1111726 Respironics.

When I start it, it appears the message E-47. And do no more. What is the problem? How can I solve it?

Log inor Register

to rate and post a comment

Chad Looney

Re: Error 47 (E-47) in the Remstar /pro2 humidifier

December 20, 2007 03:16

E-47 is a Flow Sensor defective code. On the Circuit Board, Needs replaced at circuit board level or the whole circuit board. 806-352-4140 / qualityres@amaonline.com

Log inor Register

to rate and post a comment