error code 032 on Siemens Polydoros IT 80

(32)

(1)

Henry Lopez

error code 032 on Siemens Polydoros IT 80

April 01, 2014 11:43

Hi everyone I'm getting an error 032 on a Siemens Polydoros IT 80 can anyone tell me what this means.

Thanks Henry

Log inor Register

to rate and post a comment

(183)
(38)
(1)

Dante Nuevo

re: error code 032 on Siemens Polydoros IT 80

April 01, 2014 04:44

Check your 5+VCC and adjust it 5.5 VDC.

Err032: Real Time Clock Data Error 0
Description:
• The real-time clock on the D500 contains invalid data.
Possible Causes of Error:
• 5 V VCC on the D500 board too low
• Missing configuration data
• D500 defective.
Corrective Measures:
• Adjust the low voltages (Adjustment Instructions)
• D500 reconfiguration (for data: see Operating Instructions); if there is still an error message
-> replace D500

Log inor Register

to rate and post a comment

(5)

houssein ibraheem

re: re: error code 032 on Siemens Polydoros IT 80

May 28, 2018 10:36

good day,

i have these errors 242/032 246/032 248/032
the system is sireskop cx with polydoros 50

Log inor Register

to rate and post a comment

(1)

Murat MIHLAR

re: re: error code 032 on Siemens Polydoros IT 80

October 11, 2018 08:08

hi
power supply is okey but error 0032 is continues
please help me

Log inor Register

to rate and post a comment

Mark Parris

re: error code 032 on Siemens Polydoros IT 80

April 01, 2014 04:44

Err032: Real Time Clock Data Error 3
Description:
• The real-time clock on the D500 contains invalid data.
Possible causes of the error:
• Missing configuration data
• D500 defective
Corrective measure:
• Perform reconfiguration of D500 (for data, see the Operating Instructions); if error
message persists -> replace D500 and perform reconfiguration and readjustment
(see Err003: D500 Error)

You need the manual

Log inor Register

to rate and post a comment

(1)

CHRIS ETHEN

re: error code 032 on Siemens Polydoros IT 80

April 01, 2014 04:45

Err032 reflects a problem with the real time clock. Reconfiguring your D500 board may help. If not, it may need to be replaced. Part Number is 8362175. Good luck!

Log inor Register

to rate and post a comment

Nationwide Service & Support 1-866-NUC-MED1

OEM trained nationwide service for NM, PET, PET/CT systems. Relocation, installation, parts, technical and clinical support. Contact 1-866-NUC-MED1 or info@nis-mit.com

 

advertisement

 

advertisement
advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED