Fuji Fcr Prima question for Error 270

October 12, 2016 10:12

Hello. I have the following situation: I have Error 270, but only Warning(11270). I change to show it only in Error Log. Do I need to do something? To clean the lighting guide, or calibrate S-value ?

Best Regards...

Log inor Register

to rate and post a comment

(306)
(56)

Gabriel Cool

re: Fuji Fcr Prima question for Error 270

October 13, 2016 11:02

No. Its just warning and its not important. You can disable this massage in RU settings

Log inor Register

to rate and post a comment

(109)
(9)
(1)

Svilen Popov

re: re: Fuji Fcr Prima question for Error 270

October 18, 2016 06:39

Okay, I disable it. But what is this warning? Why it appear in RU ?

Log inor Register

to rate and post a comment

(1)

Karan Nair

re: re: Fuji Fcr Prima question for Error 270

May 13, 2019 10:30

Hi i have a problem regarding error 252 in prima t reader where it is an intermittent issue,so kindly suggest me on which parts do i need to replace or what solution do i need to offer to the customer

Log inor Register

to rate and post a comment

(109)
(9)
(1)

Svilen Popov

re: re: re: Fuji Fcr Prima question for Error 270

May 15, 2019 10:27

Error 10252: [During bootup or routine processing]
The image reading error occurred.

Check the error log for other errors. If any error occurs, analyze it.
- Check the CPU board fuses. Replace the PSU27A power supply,
the PSU27B board and CPU board if the fuse blowout recurs
immediately after replacement of the fuses.
- Reseat the connectors connecting with the scanning optics unit.
- Replace the scanning optics unit.
- Replace the SED board.
- Replace the cables connecting from the SED board to the CPU
board.
- Replace the CPU board.