fuji fcr prima T error 416

March 31, 2021 11:34

Hi
Fuji fcr prima T reads 14X14 cassette normally, but it does not read 14X17 cassette and gives error 416, other fcr read the cassette with no problem.
I opened the RU and checked the operation with 14X17 cassette it goes smoothly till the ip reaches sc3 sensor the RU reverse move direction and return the ip to the cassette.
All cards fuses checked and were ok.
Any help appreciated.

Log inor Register

to rate and post a comment

(374)
(75)
(2)

juan pablo amado

re: fuji fcr prima T error 416

April 05, 2021 11:57

11416 Subscanning conveyance error
(3)
[During routine processing]
The IP transport motors (MZ1/MC1) were driven for a reading
conveyance operation, and a leading-edge detection interrupt
occurred. The IP sensor 1 (SC3) in the conveyor unit became
OPEN, but a tailing-edge detection interrupt did not occur
within the specified time.
[When a conveyance error occurs]
- Check for mixture of debris in the IP conveyance path.
- Check for abnormalities in rotation of the conveyance roller.
[When no conveyance error occurs]
- Check to make sure that the sensors normally work by means of the
monitor function of the RU PC-TOOL.
- Check the SND board fuses.
- Replace the sensors or the motors.

Log inor Register

to rate and post a comment

(62)
(7)

ghazy

re: re: fuji fcr prima T error 416

April 06, 2021 10:50

Thank you..
I did all these steps except the last one.
When doing ip erasing before starting, the ru erase 14X17 ip normally with no problem

Log inor Register

to rate and post a comment

(374)
(75)
(2)

juan pablo amado

re: re: re: fuji fcr prima T error 416

April 08, 2021 10:54

did you do a conveyance check? By RU MUTL?

Log inor Register

to rate and post a comment