GE AMX 4 PLUS Powerng up DC AC ok but not charging

Tags: Radiology - R/F, GE

(1)

David Panganai

GE AMX 4 PLUS Powerng up DC AC ok but not charging

November 25, 2019 05:52

AMX 4+ not charging after plugging it into power. The battery has 2 bars and instrument panel displaying charge battery immediately. Any ideas on what might be causing the instrument to charge?

Log inor Register

to rate and post a comment

(13)

dms

re: GE AMX 4 PLUS Powerng up DC AC ok but not charging

November 26, 2019 12:32

do you have any error message ?

thanks

regards

Log inor Register

to rate and post a comment

(4)

Shawn Broussard

re: GE AMX 4 PLUS Powerng up DC AC ok but not charging

November 27, 2019 09:59

I would first look at the battery charger board, it may be as simple as the plug into the wall outlet, have you tried to charge it through service ?

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: GE AMX 4 PLUS Powerng up DC AC ok but not charging

November 27, 2019 04:51

How old are batteries?
Can you force charge batteries through service function?
Check operation of charger board especially the plug in relays. I don't remember the identifier but check for fried contacts.
Less likely, check charger power cable, particularly male plug end. They live a particularly tough life.
Do you have a service manual?