Over 90 Total Lots Up For Auction at One Location - WA 04/08

GE CT LIGHT SPEED ULTRA 4 SLICE

(19)
(1)

Sara Canas A

GE CT LIGHT SPEED ULTRA 4 SLICE

December 07, 2017 05:51

Hi everyone, load the LFC to an Ultra Light Speed ​​during initialization OC leave this message Recon selftest detected reconstruction errors.
I need help please
Thank you

Log inor Register

to rate and post a comment

(43)
(21)

randy oswalt

re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 08, 2017 09:28

Sounds like you have a IG issue. Verify recon queues are clear.

Log inor Register

to rate and post a comment

(19)
(1)

Sara Canas A

re: re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 13, 2017 03:20

Hello, as I know it?
Thank you

Log inor Register

to rate and post a comment

(14)
(3)

Ignacio Garcia

re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 08, 2017 09:28

When Is the error?
Furing the load from cold or after the lfc and you try to boot the system?
What Is your software versión?
What kind of OC do you have?
sysmedi22@gmail.com
I Speak spanish.

Log inor Register

to rate and post a comment

(19)
(1)

Sara Canas A

re: re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 12, 2017 09:13

If are a GOC1.

Log inor Register

to rate and post a comment

(303)
(47)

CRAIG WAMBURA

re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 10, 2017 12:05

What console do you have GOC1 or GOC3?
I need this to see if we are dealing with IG/DARC or Pegasus/Motorolla.....
If its a GOC3, you should know that the IG Node is an option on this system and you can remove it without affecting your recon speeds.
But usually the system at startup will let you know if there is a problem with IGs, but if you have recon errors, I would really look at the image disk / raw data disk...or their IDE cables.

Contact me craig@telemed-technologies.com so we can solve this

2 people liked this
Log inor Register

to rate and post a comment

(19)
(1)

Sara Canas A

re: re: GE CT LIGHT SPEED ULTRA 4 SLICE

December 12, 2017 09:13

If are a GOC1.

Log inor Register

to rate and post a comment

(9)

Daniel Cazarez

re: re: GE CT LIGHT SPEED ULTRA 4 SLICE

January 19, 2019 12:13

Hi Craig, I Have a IG/DARC Westville, I got the error Recon selftest detected reconstruction errors when system is turn on. If I make an study the image recon error appears and the reconstrucion appears after the system is rebooted.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

ISS - Contact Us Today For All Your Contrast Injector Needs: 888-667-1062

Your comprehensive contrast injector solution for 10 years to include: on-site service, telephone technical support, depot repair with loaner options, parts sales, injector sales, injector install & de-install, Level 2 & beyond training.

stats
advertisement

Upcoming Auction Events

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)