GE CT Lightspeed QXI ( scanning hardware reset failed)
July 22, 2016 10:31
Dear all,
the system failed to rest the hardware and bring out The following controllers failed to report initialization status in 120 seconds: OBC.on .
ping STC and ETC are ok.
ping OBC not respond.
when the scanning hardware reset failed the gantry
display go blank. when application is down all node OBC, STC and ETC respond to ping; when application go up and scanning hardware reset failed and Gantry diplay go blank with 3 green dots, OBC not respond to ping and stc and etc are ok to ping.
I swap OBC's CPU and LSCOM with STC without any change. I also set the DIP switch for normal config when I swap I check the slip ring and also cleaning it, but the problem still same, I also bypass the data connection from stc to obc it still same any suggestion i will apreciate
to rate and post a comment
(9)
Dean Kenney
re: GE CT Lightspeed QXI ( scanning hardware reset failed)
July 23, 2016 12:21
You probably have a issue with the ccb or dcb if you can ping the obc
to rate and post a comment
picker
re: re: GE CT Lightspeed QXI ( scanning hardware reset failed)
August 08, 2016 12:46
chang ma board scanner hardware successful
to rate and post a comment
(1)
Rajesh Venewal
re: GE CT Lightspeed QXI ( scanning hardware reset failed)
July 24, 2016 11:22
First try to Changed table elevation lower limit otherwise upper limit .
to rate and post a comment
(1)
Rajesh Venewal
re: GE CT Lightspeed QXI ( scanning hardware reset failed)
July 24, 2016 11:22
If your STC and ETC is pinging then try to change table elevation lower limit or upper limit . May be it has a problem .
to rate and post a comment