I’m Getting Gantry Error .Code: 13-1002-09 error detected by:TGP(go) gantry cover switch is open and axial motor is not Powered .Please give me solution, it’s very urgent

Log inor Register

to rate and post a comment

(9)
(4)
(5)

Bruce Hull

re: GE High Speed CT ( urgent)

December 06, 2017 09:24

There are switches for both the front and back covers. They are located on the right side of the gantry. There is also two switches for the right side cover. They are all in series.

Log inor Register

to rate and post a comment

(122)
(9)
(5)

Manuel Esquer

re: GE High Speed CT ( urgent)

December 06, 2017 09:24

CHECK MICROSWITCH RIGHT SIDE COVER

Log inor Register

to rate and post a comment

(129)
(14)
(2)

Lin Dong

re: GE High Speed CT ( urgent)

December 06, 2017 09:27

Please check the Gantry cover which is not properly installed or you need to by-pass the switch when you do the service and test

Log inor Register

to rate and post a comment

(155)
(24)

CRAIG WAMBURA

re: GE High Speed CT ( urgent)

December 07, 2017 09:25

1. Check alignment between Gantry cover and Gantry Cover Switch;
2. Check gantry Cover Switch and its cabling to TGP

Log inor Register

to rate and post a comment

Call KEI for all you MRI and CT service needs (512)477-1500.

KEI is committed to customer service. Identifying the root of problems to minimize down time & cost is key. KEI is a company you can trust when planning Equipment Moves, System Purchases, & Inspections. Call or email today at info@keimedicalimaging.com. DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED