dismiss

Save the Date - Our next Clean Sweep Live Auction will be on Tuesday, September 26th at 9:30AM EST

SEARCH
Current Location:
>
>
> GE Lightspeed 8 calibration failure

GE Lightspeed 8 calibration failure

(98)
(10)

CRAIG WAMBURA

GE Lightspeed 8 calibration failure

September 11, 2017 02:58

I am faced with a problem on a Lightspeed 8.

It had a software issue with necessitated an LFC but I found that the savestate MOD was corrupted and hence could not restore system state.

I reconfigured the system manually and now trying to do a full calibration. I have successfully completed DAS Gain, Collimator and Z-slope cals but when I start the Detailed cal, the system aborts scan at Non-Bowtie Air Cals, immediately after the first exposure.
The log shows the following errors 230015013, 230015016 and 245200....

I am trying to recover the cal_db but am getting stuck here.

Log inor Register

to rate and post a comment

(175)
(22)

Karen Sarkisov

re: GE Lightspeed 8 calibration failure

September 13, 2017 09:07

Hi! you can change MOD driver, find CD driver with best quality and instal this. then reinstal SW
About errors, can you send img of this erors?

Log inor Register

to rate and post a comment

Sentient is Transforming the pre-owned Laser & IPL industry. 855.819.3781

Sentient is the premier resource for buying, selling, repairing & refurbishing pre-owned cosmetic lasers & energy based devices. Top OEM's including Cutera, Lumenis, Palomar, Alma, Syneron & Cynosure in stock now! DOTmed Certified DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED