GE LIGTHSPEED OF 16 Slices

Tags: CT, GE

(1)

Angel Colorado

GE LIGTHSPEED OF 16 Slices

July 12, 2019 10:48

I TURN IT ON AND HE SAYS WAITING FOR START OF THE SYSTEM AND DOES NOT LOAD

Log inor Register

to rate and post a comment

(114)
(15)

VENGE

re: GE LIGTHSPEED OF 16 Slices

July 14, 2019 02:56

Hi Angel!. I can help you. Please, send me an email to guery.diaz@maxialatam.com. Hablo espaƱol!

Log inor Register

to rate and post a comment

(2)
(1)

Ruben Cadima

re: GE LIGTHSPEED OF 16 Slices

July 15, 2019 10:23

At first disconnect it from the hospital network and try to boot. If that does not help try to access to the setup to see if it reads the HDs in order to make a load from cold. Yo need to be sure that you have the system, the application and the save state CDs.

Log inor Register

to rate and post a comment

(8)

Chris Flores

re: GE LIGTHSPEED OF 16 Slices

July 15, 2019 10:28

Possible Darc not booting up

Log inor Register

to rate and post a comment

(27)
(3)

ALLEN GODFREY

re: GE LIGTHSPEED OF 16 Slices

July 18, 2019 09:13

Cris may be right, the easy way to check the darc is to hook a VGA monitor onto the back of the darc and watch the boot up, it should reach a login prompt. if the darc is ok it may be software issues. As Ruben said verify all your host hard drives are present on hardware scsi buss start up. be sure you also have your options key disc. Options entered manually will restore from the state dvd, others will need to be restored from the options dvd.