GE Logiq P5 Monitor showing NOT CONNECTED

December 18, 2018 09:37

I have a GE Logiq P5 when i give power monitor showing message NOT CONNECTED. Can someone guide me what is the problem?
Thanks

1 person liked this
Log inor Register

to rate and post a comment

(2)
(1)

Daysun

re: GE Logiq P5 Monitor showing NOT CONNECTED

December 19, 2018 09:03

Hi Mohammad
Your monitor shows that message because has no signal, first open back cover of the monitor probably it will have a VGA or DVI connection unplug that connection and connect it to a laptop with a video cable,,, just to verify if the monitor is fine, if it is good,,, check video cable, pc inside the equipment.

1 person liked this
Log inor Register

to rate and post a comment

Alfedo Rosario

re: re: GE Logiq P5 Monitor showing NOT CONNECTED

October 09, 2019 09:41

After checking the monitor,if its working properly,possible cause would be the video card or graphics card of the machine