GE Pestige SI Error E8004

Tags: Radiology - R/F, GE

(8)
(1)

Volodymyr Kozlov

GE Pestige SI Error E8004

May 27, 2017 04:09

An error occurs when entering the service mode. Mph Port is Not Open. E 8004.

Any ideas? What should I do? How to solve the problem?

1 person liked this
Log inor Register

to rate and post a comment

(18)
(4)

Nikolay Bespalow Nikolay

re: GE Pestige SI Error E8004

May 28, 2017 12:21

Replace the timecamp ??

1 person liked this
Log inor Register

to rate and post a comment

(8)
(1)

Volodymyr Kozlov

re: re: GE Pestige SI Error E8004

June 02, 2017 07:58

I changed the memory chip (Static RAM with a battery) in the generator. There was an error 3502 "Checksum RAM (PU parameters) corrupted". How to fix this?

Log inor Register

to rate and post a comment

(41)
(6)

Francesco Didonato

re: GE Pestige SI Error E8004

May 28, 2017 12:22

Hello your system has GPX CONSOLE or NORMALLY CONSOLE? If You have GPX CONSOLE You must connect your cable in the GPX and open comunication by touch screen where the is logo GE You connect in this mode?

1 person liked this
Log inor Register

to rate and post a comment

(8)
(1)

Volodymyr Kozlov

re: re: GE Pestige SI Error E8004

June 02, 2017 07:58

I have a GPX CONSOLE. I changed the memory chip (Static RAM with a battery) in the generator. There was an error 3502 "Checksum RAM (PU parameters) corrupted". How to fix this? From the touch panel I go into the service mode. The console setup works, the communication test is performed normally, and the MPH setting writes an error E8004 "MPH port not opened or installed".

Log inor Register

to rate and post a comment

(8)
(1)

Volodymyr Kozlov

re: GE Pestige SI Error E8004

June 02, 2017 07:59

I have a GPX CONSOLE. I changed the memory chip (Static RAM with a battery) in the generator. There was an error 3502 "Checksum RAM (PU parameters) corrupted". How to fix this? From the touch panel I go into the service mode. The console setup works, the communication test is performed normally, and the MPH setting writes an error E8004 "MPH port not opened or installed".

Log inor Register

to rate and post a comment

(41)
(6)

Francesco Didonato

re: re: GE Pestige SI Error E8004

June 04, 2017 01:46

I hope You are One backup for restore parameter otherwise You must ti recalibrate all' generator so error 3502 is checksum error You must ti ho in menù 5 select checksum and touch F1 and V togheter

1 person liked this
Log inor Register

to rate and post a comment

(8)
(1)

Volodymyr Kozlov

re: re: re: GE Pestige SI Error E8004

June 05, 2017 03:46

The settings are not saved. How to enter the menu number 5 in the GPX SI console? It is not on the screen with the touch panel.

Log inor Register

to rate and post a comment

(41)
(6)

Francesco Didonato

re: re: re: re: GE Pestige SI Error E8004

June 05, 2017 04:57

For entering in the generator You need ti connect with laptop lower GPX console with cable null modem You must ti have software for MPH generator installed on your PC but i'm afraid thai You don't have backup it's problem if You don't know the system

Log inor Register

to rate and post a comment

(8)
(1)

Volodymyr Kozlov

re: re: re: re: re: GE Pestige SI Error E8004

June 06, 2017 10:28

Can you help with the software for the tuning laptop? My mailbox is Volodymyr.kozlov@gmail.com

Log inor Register

to rate and post a comment

(8)
(4)

Amirbayov Zohrab

re: GE Pestige SI Error E8004

February 21, 2019 10:26

Have you solved your problem? I have a lot of material on the system Prestige II. these systems are similar. I myself am looking for firmware for HDD3.21 GPX console. or rather, drivers for the touchscreen. my address is
zohraba661@gmail.com

Log inor Register

to rate and post a comment

(8)
(4)

Amirbayov Zohrab

re: GE Pestige SI Error E8004

February 21, 2019 10:26

Have you solved your problem? I have a lot of material on the system Prestige II. these systems are similar. I myself am looking for firmware for HDD3.21 GPX console. or rather, drivers for the touchscreen. my address is
zohraba661@gmail.com

Log inor Register

to rate and post a comment

RaySafe helps you avoid unnecessary radiation

RaySafe solutions are designed to minimize the need for user interaction, bringing unprecedented simplicity & usability to the X-ray room. We're committed to establishing a radiation safety culture wherever technicians & medical staff encounter radiation.

 

advertisement

 

 

advertisement

 

advertisement
Advertisement
Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2019 DOTmed.com, Inc.
ALL RIGHTS RESERVED