Ge Senographe Dmr E12 Checksum Error 07181 - Service Technicians Forum

Cleansweep Live Auction Thursday Jan. 21th 10AM EST - Pre-Bid Now
California Auction Event Closing Jan. 26th - Bid Now

GE SENOGRAPHE DMR E12 Checksum error 071/81

(52)
(8)

Tatiana Bautista

GE SENOGRAPHE DMR E12 Checksum error 071/81

January 11, 2021 10:38

we have a problem with the mamofrafo dmr ge, in past days appeared the following error E12 Checksum error 071/81, we already changed the batteries of 3.6, we have pc with serial port rs232, but we do not have the software mamocom , can help me to solve our problem thanks in advance

Log inor Register

to rate and post a comment

(33)
(2)

Carlos Pons

re: GE SENOGRAPHE DMR E12 Checksum error 071/81

January 13, 2021 09:55

Beside the mamocom program you will need a good backup from the system in order to restore all the information. Otherwise you have to recalibrate the whole system.

Log inor Register

to rate and post a comment

(52)
(8)

Tatiana Bautista

re: re: GE SENOGRAPHE DMR E12 Checksum error 071/81

January 13, 2021 11:22

Hi, we have good backup, Can you help me with the correct procedure to upload the backup to the mammograph, should I put the mammograph in service mode?

Log inor Register

to rate and post a comment

Your Largest Midwest Source for Philips & GE Patient Monitoring

Gopher Medical specializes in Philips & GE Patient Monitors, Modules & Telemetry. All IntelliVue Monitors, MMS, M3015A, M4841A, MX40, PRO CH, PRO FH, GE B450, B650, SAM, Dash Monitors, REPAIRS, PARTS, & SERVICE, Call 844-246-7437 sales@gophermedical.com

DOTmed Certified DOTmed 100
advertisement

 

RSTI Newsletter

 

Hitachi

Upcoming Auction Events

ZRG LLC.
TIMED
Jan 26, 2021 (99 lots)