I have this situation with Silhouette vr:
The equipment starts normally and complete the starting procedure, complete the ready (tube rotates normally) but the ready lamp does not lit and when pressing Exposure button the exposure signal lasts for a very short time (the exposure is interrupted) no matter what was the time or mAs. (There is the exposure beep but for a very short time). No error massage displayed.
Any help will be appreciated.

Log inor Register

to rate and post a comment

(22)
(1)

Martin Martinez

re: Ge Silhouette vr cuts exposure

September 07, 2017 10:35

Hello Ghazy.
You have to test the input power in each HV Tank and if the input is not ,check the inverter, if the input is ok, then test the output of each HV Tank.
Be carfull.

Log inor Register

to rate and post a comment

(48)
(5)

ghazy

re: re: Ge Silhouette vr cuts exposure

September 13, 2017 09:07

Thank you Martin I'll check and inform you the results.

Log inor Register

to rate and post a comment

(48)
(5)

ghazy

re: re: Ge Silhouette vr cuts exposure

September 21, 2017 09:00

It was the NVRam battery, thank you

1 person liked this
Log inor Register

to rate and post a comment

(26)
(2)

Carlos Pons

re: Ge Silhouette vr cuts exposure

September 08, 2017 10:17

Hello Ghazy
You can go to service under pc and do some tests, rotation, Hv no load filament etc, How old is the system? it could be the NVram battery in the Kv control..

Log inor Register

to rate and post a comment

(48)
(5)

ghazy

re: re: Ge Silhouette vr cuts exposure

September 13, 2017 09:07

Hello Carlos, Thank you for replay, its installed in 2012, i'll do these tests and inform you.

Log inor Register

to rate and post a comment

(48)
(5)

ghazy

re: re: re: Ge Silhouette vr cuts exposure

September 21, 2017 09:00

You were right it was the NVRam battery. I replaced it and Download the backup And the system is fine now. Thanks.

Log inor Register

to rate and post a comment

Diagnostic Imaging Training: RSTI (Radiological Service Training Institute)

15,000 service professionals from over 50 different countries have trained in over 35 courses in radiology, mammography, CT, ultrasound, networking, PACS, & DICOM. Classroom & lab facilities to accommodate up to 60 students. DOTmed 100

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED