GE Vivid i Freezing Issues

January 23, 2020 11:01

Greeting
Im working on a Vivid i with freezing issues. I think that the problem is caused by the BEP assy. I tried to perform a software reload and the system still with the same problem in intermittent times. I swap the HDD to try again the software reload and the system was freezes in the middle of the base installation, as the original HDD. Then, I remove the ribbon cable between the BEP and RFI and tried the software reload. Nothing change. If I replace the BEP, the original system option key could be change or the original serial key should be works fine with the new BEP assy?
Thanks in advance for your help
Have a nice day

Log inor Register

to rate and post a comment

(16)
(2)

Roman Huda

re: GE Vivid i Freezing Issues

January 31, 2020 10:20

Hello.

Option key saved in software.

My advice - open machine, remove etx/xtx module and clean, maybe need replace thermal interface (like rubber) prepare new hdd, upgrade BIOS and install new software.
Maximum software - BT12

Log inor Register

to rate and post a comment

(37)
(2)

Jesus E Mendez

re: re: GE Vivid i Freezing Issues

February 03, 2020 09:43

Ok, Thanks for answer my post. I will try this step. Have a nice day and thank you very much for your help

Log inor Register

to rate and post a comment

(10)

Cali Mohamed

re: GE Vivid i Freezing Issues

January 31, 2020 10:22

Hello, befor changing the bep, reload the software in a new HDD, sometimes the HDD makes that kind of problem ALSO. CHANGE THE MEMORY

Log inor Register

to rate and post a comment

(77)
(17)

Brian Eberling

re: GE Vivid i Freezing Issues

February 19, 2020 10:29

If you change the BEP, you will have to reprogram your system ID from the menu on your software installation media.

Log inor Register

to rate and post a comment

Auction Alert - Live Now: ZRG Medical Surplus Auctions March 24th

ZRG Medical closeout auctions take place on DOTmed.com. Bid today on 75 lots of medical equipment. Onsite or virtual inspection available. All equipment is Biomed tested with detailed description. Sold As-Is to highest bidder.

DOTmed Certified DOTmed 100
advertisement

Upcoming Auction Events

ZRG Medical
TIMED Ending Soon
Mar 24, 2023 (90 lots)
Surplus Service
TIMED
Mar 27, 2023 (71 lots)