GE X-RAY GENERATOR EXPOSURE ERROR

Tags: PACS / DR / CR / Digital, GE

(7)

George Anyetei Otchere

GE X-RAY GENERATOR EXPOSURE ERROR

March 20, 2017 10:33

My GE X-ray is giving error and can not run.
"0*0060 generator exposure error has occured. Release exposure switch and try again."

I will really appreciate it if there are any suggested solutions that can help. Thanks.

Log inor Register

to rate and post a comment

(2260)
(430)
(12)

Lenny Place

re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 10:19

Which generator has this error I am not familiar with a error code like this

Log inor Register

to rate and post a comment

(7)

George Anyetei Otchere

re: re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 04:01

IT DOES NOT SPECIFY IT ONLY SAYS GENERATOR EXPOSURE ERROR HAS OCCURRED

Log inor Register

to rate and post a comment

(6)
(2)
(9)

Eldad Tor

re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 10:20

Hi, i can help you.
dean@mtcmed.com

Log inor Register

to rate and post a comment

(45)
(11)

Kamran Butt

re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 10:20

You have to share Model of system for getting response.

Log inor Register

to rate and post a comment

(7)

George Anyetei Otchere

re: re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 04:01

MODEL IS : DR-F/BRIVO XR385

Log inor Register

to rate and post a comment

(98)
(11)

CRAIG WAMBURA

re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 10:27

What GE Model is this and do you have the full error code?

I would troubleshoot my HV section including HV Cables and Tank.
Is your mains power supply ok?
Could also be a faulty exposure switch

Log inor Register

to rate and post a comment

(7)

George Anyetei Otchere

re: re: GE X-RAY GENERATOR EXPOSURE ERROR

March 21, 2017 04:01

MODEL; DR-F/BRIVO XR385
ERROR CODE IS 0*0060
I CHECKED CONTINUITY FOR THE HT CABLES AND OK. WHAT SHOULD I CHECK ON THE TANK. MAIN POWER SUPPLY OK

Log inor Register

to rate and post a comment

Nationwide Service & Support 1-866-NUC-MED1

OEM trained nationwide service for NM, PET, PET/CT systems. Relocation, installation, parts, technical and clinical support. Contact us at info@nis-mit.com

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2017 DOTmed.com, Inc.
ALL RIGHTS RESERVED