hello friends I need help with a vivid IQ, the power supply

September 07, 2021 02:33

hello friends I need help with a vivid IQ, the power supply has been damaged, I have changed it 3 times and I can not find the damage. It works a few days and it will be damaged again.

Log inor Register

to rate and post a comment

(165)
(23)

Tibor Ivanszky

re: hello friends I need help with a vivid IQ, the power supply

September 08, 2021 11:00

Do you think the enviroment specifications are met with requirements? Is the room temperature, humudity are ok? Dust filters are cleaned? UPS ( uninterruptible power supply) installed?

Log inor Register

to rate and post a comment

(3)

RUSTECH

re: hello friends I need help with a vivid IQ, the power supply

September 09, 2021 10:56

Hello! Is it the same power supply or 3 different?

Log inor Register

to rate and post a comment

(2)

sergio eduardo figueroa arias

re: re: hello friends I need help with a vivid IQ, the power supply

September 10, 2021 10:51

they are 3 different

Log inor Register

to rate and post a comment

(3)

RUSTECH

re: re: re: hello friends I need help with a vivid IQ, the power supply

September 13, 2021 10:48

Then we would suppose you have got an issues with power surges in your electricity network. Almost all GE's power supplies are very weak (and we've got idea why).
You could burn them out easily if you have no a power stabilizer at least @ 3 KWT.
As for GE Vivid power supply - it could be repaired easily in 50% of cases.
Usually the power surge burns first semiconductors and capacitors in power chain.

Log inor Register

to rate and post a comment

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

advertisement

 

RSTI Newsletter

 

Hitachi

Upcoming Auction Events

USEDSYNTHES.COM
TIMED Ending Soon
Sep 28, 2021 (44 lots)
Premier Medical Depot
TIMED
Sep 29, 2021 (979 lots)
ZRG Medical
TIMED
Oct 01, 2021 (120 lots)
Surgamed
TIMED
Oct 05, 2021 (310 lots)