Hi- Wondering if anyone can help advise us, we have a Fuji CR XG-1 with a stuck cassette. Normally we can remove it after resetting the unit, but now it reports "SYSTEMDOWN" with the yellow exclamation point lit and doesn't seem to go through its process to eject the cassette. Telnet and ping work, so the unit appears to be on the network properly; any advice for what I might do to see if the unit requires repair or another method to reset it? Thanks in advance!

Log inor Register

to rate and post a comment

(34)
(7)

xtekvet

re: Help with Fuji CR XG-1 stuck cassette

December 05, 2014 10:19

Have you tried the manual release. It is a small hole on the right side of the unit as you are facing its front. It is about the level of the cassette tray. You can use a large , bent paper clip.

Log inor Register

to rate and post a comment

(1)

daniel ace enrijo

re: re: Help with Fuji CR XG-1 stuck cassette

March 19, 2020 10:44

Hi, my machine is FCR prima t2 and have the same problem. the cassette is stuck and error code 422 and all lights blinking. is there any chance i could fix this without calling a technician? thanks in advance

Log inor Register

to rate and post a comment

(3)

Naeem Ahmed

re: Help with Fuji CR XG-1 stuck cassette

July 17, 2019 09:41

There are some reasons to clear it. First tell me all sizes are stucking or only one particular size stucking. ?

General Solution
1. Cassette eject system required service.
2. Replace the cassette.