Hvdc Errors 183182 And 214080 - Service Technicians Forum
advertisement
Current Location:
>
>
>
> HVDC errors 183182 and 214080

HVDC errors 183182 and 214080

Tags: CT

(4)

Ali AlZein

HVDC errors 183182 and 214080

August 11, 2019 02:25

I am getting an error 183182 on Tube Warm up

host:obc error:183182
i/o board pilot relay are closed. PDU pilot relay not energized. BCSTAT=0; HVDC rail voltage=215v.

and getting an error 214080 upon KV & mA test

host:obc error:214080
HVDC BUS Voltage is present while backup contactor pilot relay indicates a de-energized state, gentry i/o relay k251 and k245 are energized
hvdc bus voltage:210 V
gentry i/o address:FFAFBFH, bit:D4 expected 1 actual :0


Please guide me through the right Troubleshooting Flowchart to figure out the root cause of these errors.

Upon turning off the power at the PDU, I tested the resistance across A2TS1 cables at the PDU and these were the readings:
Red to Ground 53 K ohm
Black to Ground 50 K ohm
Black to Red 451 Ohms
I tested the phase to phase voltage at the input to the PDU and all read around 380VAC, and phase to ground around 215VAC.

Your help is much appreciated.

waiting for your kind reply

Best Regards,
Ali AlZein

Log inor Register

to rate and post a comment

(4)

Ali AlZein

re: HVDC errors 183182 and 214080

August 13, 2019 10:42

Sorry forgot to mention, I am dealing with GE CT Lightspeed LSX.3, performix tube.

Log inor Register

to rate and post a comment

(4)

Ali AlZein

re: re: HVDC errors 183182 and 214080

August 14, 2019 10:09

DS5 MN_FLT red LED is lit on the CPDU

Log inor Register

to rate and post a comment

OiHealthcare - ISO 13485 Certified - Call 1-888-278-9933

Oxford Instruments Healthcare, is a leading ISO 13485-certified organization, that specializes in providing quality refurbished CT & MRI systems, service, mobile units & parts at prices you can afford. Call 888-278-9933 Email DOTmedinfo@oxinst.com

 

advertisement

 

advertisement

 

advertisement

 

advertisement
advertisement