Karl Storz Endomat Hamou trouble

(59)
(23)

Yuriy Gontarenko

Karl Storz Endomat Hamou trouble

October 26, 2011 06:17

Karl Storz Endomat Hamou could not pass POST (power on self test).
After the turning on just 5 blinkings from the front panel, that's is.
Does anybody have an error codes list or service manual to it ?

Log inor Register

to rate and post a comment

(8)
(3)

Victor Kolchenko

Re: Karl Storz Endomat Hamou trouble

October 27, 2011 10:12

Error codes
If an error occurs when the device is switched on, an intermittent warning tone lasting approx. 10
seconds sounds. The error display continues to flash until the error is remedied. The error code is
indicated by the number of flashes.
Only one error code, namely the first to be detected, is displayed at any one time. In addition, the error
code is also shown on the “Actual flow” display. The error code corresponds to the number of LEDs
which are lit up.

err code 5
Offset error in irrigation pressure sensor; measured pressure

1 person liked this
Log inor Register

to rate and post a comment

(5)

christophe dubernard

re: Re: Karl Storz Endomat Hamou trouble

October 16, 2013 08:22

hi
I'm looking for a service manual of this endomat (.pdf) , can you help me pls
tanks

Log inor Register

to rate and post a comment

(4)

Mesystems-Ukraine

re: Re: Karl Storz Endomat Hamou trouble

January 15, 2017 07:39

Hi Victor, We have the same problem. Help me pls

Log inor Register

to rate and post a comment

(8)
(3)

Victor Kolchenko

re: re: Re: Karl Storz Endomat Hamou trouble

January 23, 2017 08:00

Hi,
you can contact me by vkolch@email.cz for further details.

Log inor Register

to rate and post a comment

(2)

ramin eleskerov

re: Re: Karl Storz Endomat Hamou trouble

January 12, 2018 09:38

Hi i have problem Storz endomat 263310
indicates error 11 when the device is connected. I do not have a service manual. You can help me?

Log inor Register

to rate and post a comment

(1)

(2)

Jose Francisco Cervantes M

re: Karl Storz Endomat Hamou trouble

November 29, 2012 08:56

Hi yuriy, did you solve the problem whit the hamou endomat?
I have one that make almost the same, but it keeps beeping and blinking.

Log inor Register

to rate and post a comment

(1)

Fernando silva

re: Karl Storz Endomat Hamou trouble

August 22, 2015 10:24

Hello how are you? Where can I find the pressure sensor , which is adjacent to the optical fiber sensor. Which fits the hose sensor.

Log inor Register

to rate and post a comment

Your trusted and knowledgeable source for Medical Imaging Parts & Equipment

Dedicated Solutions-We provide dependable, cost effective solutions that exceed our customer's expectations & offer high quality services for your specific needs-Repairs, CT and MRI parts, staging, 24/7 Tech support, & same day shipment. Call 844-293-2057 DOTmed 100

 

 

 

Access and use of this site is subject to the terms and conditions of our LEGAL NOTICE & PRIVACY NOTICE
Property of and Proprietary to DOTmed.com, Inc. Copyright ©2001-2018 DOTmed.com, Inc.
ALL RIGHTS RESERVED