nikolai pavlov

Kodak 2100 x-ray Error 13

April 19, 2016 05:21

My kodak 2100 x-ray is giving me Error 13
I can't take x-ray

Log inor Register

to rate and post a comment

(13)
(6)

Pavel Kurkov

re: Kodak 2100 x-ray Error 13

April 21, 2016 08:53

Hi!Check the resistance btwn the big blue wire and the small yellow of the white connector. It must be like 12-14K. If not, check it directly on the tubehead...
Quite possible you have the ceramic divider damaged or wires unsoldered inside the tubehead

Pavel

1 person liked this
Log inor Register

to rate and post a comment

(7)

Gerson Borrero

re: re: Kodak 2100 x-ray Error 13

December 02, 2016 02:19

hi

Log inor Register

to rate and post a comment

(6)

evan wasserman

re: re: Kodak 2100 x-ray Error 13

December 28, 2017 11:52

I measure 88K between small yellow and big blue
Open up tube head and get same measurement there.
should I disconnect from board before measurement?
other place says 11-15 ohms between yellow and blue??
more details please.

wires go to white round disk which looks ok and is sealed around the outside.

how can I get manual...??

Log inor Register

to rate and post a comment

(1)

Danie Roux

re: re: Kodak 2100 x-ray Error 13

September 13, 2021 10:45

Good day. I am getting 12.4K reading between Big Blur and small Yellow. What is the next step.

Log inor Register

to rate and post a comment

(3)

nikolai pavlov

re: Kodak 2100 x-ray Error 13

April 22, 2016 09:32

Thank you Pavel. Will check it out.
Very nice of you.

Log inor Register

to rate and post a comment

(3)

nikolai pavlov

re: re: Kodak 2100 x-ray Error 13

April 27, 2016 06:46

Hi Pavel
The maintenance guy check the x-ray machine and told me is possible the replace the all board.I show him your message.
What do you think?
Thank you for your help.
PS. Pavel I think you are Russian.Are you?
My father name is Pavel Pavlov and I am Bulgarian.

Log inor Register

to rate and post a comment

(13)
(6)

Pavel Kurkov

re: re: re: Kodak 2100 x-ray Error 13

May 03, 2016 09:56

I repaired 8 tubeheads of Kodak 2X00 and the boards have never broken down. Has maintenance service checked the resistance of high voltage divider I told you last time?
PS. Yep, Im russian. I'm not sure if I can write here in other language except for english. If you got any questions feel free to email me: kurkovpavel@gmail.com

2 people liked this
Log inor Register

to rate and post a comment

jose Pinto

re: re: re: re: Kodak 2100 x-ray Error 13

September 30, 2016 05:28

Good evening. I have one problem in x-ray kodak 2100 with error 13. It is possible please help? thank you

Log inor Register

to rate and post a comment

(7)

Gerson Borrero

re: re: re: re: re: Kodak 2100 x-ray Error 13

December 02, 2016 02:19

Hi Jose, you are from? I´m from Venezuela, my email libano_987@hotmail.com

Log inor Register

to rate and post a comment

(7)

Gerson Borrero

re: re: re: re: Kodak 2100 x-ray Error 13

December 02, 2016 09:17

Hi Mr. Pavel, The internal weld is tin ?

Log inor Register

to rate and post a comment

(1)

Riedwaan Cassiem

re: re: re: re: Kodak 2100 x-ray Error 13

February 27, 2020 09:39

Hi Pavel, any info on taking he xray head apart to repair the wires if it is loose inside the head

Log inor Register

to rate and post a comment

Servicing GE/Siemens Nuclear Medicine equipment with OEM trained engineers

Numed, a well established company in business since 1975 provides a wide range of service options including time & material service, PM only contracts, full service contracts, labor only contracts & system relocation. Call 800 96 Numed for more info.

advertisement

 

RSTI Newsletter

 

Hitachi

Upcoming Auction Events

USEDSYNTHES.COM
TIMED Ending Soon
Sep 28, 2021 (44 lots)
Premier Medical Depot
TIMED
Sep 29, 2021 (979 lots)
ZRG Medical
TIMED
Oct 01, 2021 (120 lots)
Surgamed
TIMED
Oct 05, 2021 (310 lots)